GE MPPU Error 4020

February 21, 2013 12:28

Does anyone have any solutions for a 4020 error. This error comes up after console boots. Console boots with normal display and selections can be made, but prep and exposure are not possible. Tried replacing ASC and ECM CPU's & TIRC, with no success. Thank you... reply to: michael@remestamedical.com

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: GE MPPU Error 4020

February 21, 2013 10:43

4020 is generic means not communicating with node

Log inor Register

to rate and post a comment

Steven Roberts

re: GE MPPU Error 4020

February 22, 2013 12:37

Have you checked the CPU on the MPPU? See if you have an error light on the MPPU CPU board. The lights should be green. Also if you need to reset the CPU, you have to remove it from the chassis then remove the battery (will be ty wrapped to board) then short out the terminals of the 3test points protuding from the CPU while the battery is out and with the board removed from the chassis.
If this does not solve the problem, you will need a new CPU.
Hope this helps