Hi:

On the diagnostic mode for GE MPPU generator, I get error 84A3 on the power unit. Does anyone know what this error code is?
Thank you,
Dinu

Log inor Register

to rate and post a comment

(244)
(19)

Staate Hayward

Re: GE MPPU Error 84A3

December 03, 2010 03:17

Mppu Control to Mppu power unit communications check. Checks that an exposure can be initiated from the ECM (Advantx). Verifies feedback from the high voltage bleeder. Check comm cables and all connections to/from the Advantx, between the control and power cabinets, to the high voltage transformer.

Log inor Register

to rate and post a comment

(34)
(5)
(1)

Dinu Papoiu

Re: GE MPPU Error 84A3

December 05, 2010 09:38

Thank you very much.
I really appreciate your help.It clarified a lot of questions which I had about the exposure problem.