Mppu Control to Mppu power unit communications check. Checks that an exposure can be initiated from the ECM (Advantx). Verifies feedback from the high voltage bleeder. Check comm cables and all connections to/from the Advantx, between the control and power cabinets, to the high voltage transformer.
From preventive maintenance through comprehensive cryogen services, BMI-PMI is your trusted partner for new/reconditioned sales, installation, and service of both facility-based and mobile MRI solutions. www.brownsmedicalimaging.com / 800-701-9729
Dinu Papoiu
GE MPPU Error 84A3
November 29, 2010 10:55
Hi:
On the diagnostic mode for GE MPPU generator, I get error 84A3 on the power unit. Does anyone know what this error code is?
Thank you,
Dinu
to rate and post a comment
Staate Hayward
Re: GE MPPU Error 84A3
December 03, 2010 03:17
Mppu Control to Mppu power unit communications check. Checks that an exposure can be initiated from the ECM (Advantx). Verifies feedback from the high voltage bleeder. Check comm cables and all connections to/from the Advantx, between the control and power cabinets, to the high voltage transformer.
to rate and post a comment
Dinu Papoiu
Re: GE MPPU Error 84A3
December 05, 2010 09:38
Thank you very much.
I really appreciate your help.It clarified a lot of questions which I had about the exposure problem.
Dinu