Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

GE MVP 60 Error Code 172

(2)

(3)

Kevin Masterson

GE MVP 60 Error Code 172

August 17, 2011 02:51

I need help identifying this error code. Can anyone direct me to a manual listing error codes? thanks

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

Re: GE MVP 60 Error Code 172

August 17, 2011 03:47

if you send me a email address I will send a attachment with the error code.

Log inor Register

to rate and post a comment

(21)
(2)
(3)

John Vanderwerf

Re: GE MVP 60 Error Code 172

December 14, 2011 03:57

Hello Lenny, I'm looking for cal. procedure for mvp, have box but no instructions, thanks, John.

Log inor Register

to rate and post a comment

(1)

Randy Cornejo

re: Re: GE MVP 60 Error Code 172

February 18, 2014 01:20

Hi John, not sure if your still active here but I was wondering if you might have a GE part# for the cal box. Trying to locate one but without the part number it's proving difficult.

Log inor Register

to rate and post a comment

(21)
(2)
(3)

John Vanderwerf

Re: GE MVP 60 Error Code 172

December 14, 2011 03:57

Hello Lenny, I'm looking for cal. procedure for mvp, have box but no instructions, thanks, John. jvanderwerf@aol.com

Log inor Register

to rate and post a comment

(244)
(19)

Staate Hayward

Re: GE MVP 60 Error Code 172

August 19, 2011 12:44

From the Book of GE
172 CLLOOP
1st a determined period of time has elapsed after the exp command was sent to the power unit and no exposure has been initiated.
2nd CLLOOP signal detected from the power unit without any prior exp command.

Log inor Register

to rate and post a comment

(6)
(1)

Bob Buaas

re: Re: GE MVP 60 Error Code 172

January 03, 2017 10:17

If by any chance you have a pdf copy of the service manual, could you email it to me? Thanks! Bob Buaas xrayservice@comcast.net