Dear colleagues,I have a g e mvp micro >We started the install and have no service books for it.If anybody can help it would be greatly appreciated.
We boot but have no rotor,locks,and tech overload on small focus
The first two thoughts are the door interlock for the room and tube pressure swith for the rotor. You either need to connect them or jumper them out. Sorry don't have any books. Just to be sure we are talking aobut a Micro. MVP is a 3 phase unit and MVP Micro is a single phase unit. Check the small focal spot and and high voltage cable just to be sure that they are good.
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
doug westlake
ge mvp micro
May 20, 2013 06:07
Dear colleagues,I have a g e mvp micro >We started the install and have no service books for it.If anybody can help it would be greatly appreciated.
We boot but have no rotor,locks,and tech overload on small focus
PLEASE HELP!
E-MAIL contrastimagingservices@yahoo.com
to rate and post a comment
Kim Schofield
re: ge mvp micro
May 21, 2013 09:54
http://apps.gehealthcare.com/servlet/ClientServlet?REQ=RNEW&MODALITY=XR
Select MVP
to rate and post a comment
Tom Walker
re: ge mvp micro
May 22, 2013 10:05
The first two thoughts are the door interlock for the room and tube pressure swith for the rotor. You either need to connect them or jumper them out. Sorry don't have any books. Just to be sure we are talking aobut a Micro. MVP is a 3 phase unit and MVP Micro is a single phase unit. Check the small focal spot and and high voltage cable just to be sure that they are good.
to rate and post a comment
doug westlake
re: re: ge mvp micro
May 23, 2013 10:13
Thank you for the help i have yhe system up but no collimation or locks.
to rate and post a comment
doug westlake
re: ge mvp micro
May 22, 2013 10:05
it cannot be found in the library,mvp is there but no results under any category
to rate and post a comment
Manuel Casanova
re: ge mvp micro
May 22, 2013 08:53
CHECK YOUR E MAIL....
to rate and post a comment