Can some body please help on this with GE precision 500D
Getting the following errors.
Error Code
Error Text
Subsystem
3600106
Time out while expecting CAN message from generator
SysCntrl
5200118
System failed to complete boot
IUI
5800010
User time out on Boot
SysCntrl
3600024
Jedi power up failure
SysCntrl
3600029
XRTechMgr failed to get Notify Jediid from Jedi
SysCntrl
5800020
The UPS battery voltage is low
SysCntrl
Note: Voltage measurements in the “Low Voltage Power supply board” indicates no Output voltages in the Main Inverter section. This may indicate that the Power supply board is defective or signals entering the board are not correct.
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
Hussein Ali
GE Precision 500D problem
September 05, 2012 01:58
Can some body please help on this with GE precision 500D
Getting the following errors.
Error Code
Error Text
Subsystem
3600106
Time out while expecting CAN message from generator
SysCntrl
5200118
System failed to complete boot
IUI
5800010
User time out on Boot
SysCntrl
3600024
Jedi power up failure
SysCntrl
3600029
XRTechMgr failed to get Notify Jediid from Jedi
SysCntrl
5800020
The UPS battery voltage is low
SysCntrl
Note: Voltage measurements in the “Low Voltage Power supply board” indicates no Output voltages in the Main Inverter section. This may indicate that the Power supply board is defective or signals entering the board are not correct.
to rate and post a comment
Juan de la Cruz Martinez
re: GE Precision 500D problem
September 05, 2012 06:20
What is the first error that comes up, normally all other are consequence of this first error.
to rate and post a comment
Hussein Ali
re: re: GE Precision 500D problem
September 06, 2012 04:51
The error order is as it is un above in above list
to rate and post a comment
mike lozito
re: GE Precision 500D problem
September 06, 2012 02:00
Sounds like the UPS is not working properly anymore. remember its just a short term battery back up, and its batteris go bad
to rate and post a comment
Hussein Ali
re: re: GE Precision 500D problem
September 06, 2012 04:56
Will check that .. Thank u guys
to rate and post a comment
Hussein Ali
re: re: GE Precision 500D problem
September 17, 2012 05:47
UPS is fixed, but problem still exists ? any hint ?
to rate and post a comment