Over 500 Total Lots Up For Auction at One Location - NJ 06/15

GE Precision RXi

Kirill Dyakov

GE Precision RXi

December 22, 2015 04:08

Precision RXi-65
During the work sometimes appears the Error 36(+15V). If someone has error list with their descriptions and service manual, send me please to panasonic65@mail.ru
Here is photo of the error
http://s015.radikal.ru/i331/1512/b1/d7a5468a558b.jpg

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: GE Precision RXi

December 23, 2015 02:35

I will send you the troubleshooting manual error 36 will refer you to error 31 power supply issue