Hello
I have GE Prodigy densitometry. Recently the'm installing, but it happens that when you finish installing the Limits Switchs not arrive in position. I have the service manual, but it does not talk about how to adjust the switch or a image to give me an idea. Anyone have a documentation about it?
My email is: doscarperezwiccan@hotmail.com
Regards
to rate and post a comment
Desmond Johnson
re: GE Prodigy Densitometer Installation Manual
October 02, 2013 08:33
All the alignment\test procedures are in the D:\Applications\Program Files\enCore\ . The switch setting requires a Home jig to set the beginning point.
to rate and post a comment
Doscar Ernesto Perez Pereyra
re: re: GE Prodigy Densitometer Installation Manual
October 02, 2013 09:20
i solved the problem. Thanks for help.
to rate and post a comment