I have a GE Prodigy that gives me a Detector Command Failed error every time I do a QA. Any help or suggestions would be greatly appreciated.
Thanks
j.mendoza@scmes.com
Reseat the connectors on the CSBC and Detector motherboard. There is also a connector in the wiring bundle where the upper and lower bundles are going to the tube and detector that causes issues
That bad part about these systems GE uses a power supply that is not adjustable and the slightest change in resistance on a connector the supply drops below 4.85 which locks up the system. I have replaced the supply with on which I can adjust up to 5.15 or you have to trace the supply wires and find out where it is dropping and repair it
Check the Error Log for System Reset Errors. If you have these move the arm to the center of the table, and the Source Tube to the center of the X axis and try again. It could be a cable. Also look at the molex on the low voltage supply. It gets flakey.
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
John Mendoza
GE Prodigy "Detector Command Failed"
May 14, 2014 12:06
I have a GE Prodigy that gives me a Detector Command Failed error every time I do a QA. Any help or suggestions would be greatly appreciated.
Thanks
j.mendoza@scmes.com
to rate and post a comment
Lenny Place
re: GE Prodigy Detector Command Failed
May 15, 2014 08:04
measure the power supplies on the detector motherboard. many times the +5 vdc supply will be low
to rate and post a comment
Tran Tuan
re: re: GE Prodigy Detector Command Failed
June 08, 2017 10:42
I check the voltage at TP38 and TP43 is 4.773VDC, Any ideas? thank you!
to rate and post a comment
Lenny Place
re: re: re: GE Prodigy Detector Command Failed
June 08, 2017 11:03
Reseat the connectors on the CSBC and Detector motherboard. There is also a connector in the wiring bundle where the upper and lower bundles are going to the tube and detector that causes issues
to rate and post a comment
Tran Tuan
re: re: re: re: GE Prodigy Detector Command Failed
June 09, 2017 08:22
I reseat the connectors between csbc and DIB board already, but problem can not fix!
to rate and post a comment
Lenny Place
re: re: re: re: re: GE Prodigy Detector Command Failed
June 09, 2017 03:24
What is the power supply putting out?
to rate and post a comment
Tran Tuan
re: re: re: re: re: re: GE Prodigy Detector Command Failed
June 13, 2017 09:34
I measure the power supply is 4.995VDC.
to rate and post a comment
Lenny Place
re: re: re: re: re: re: re: GE Prodigy Detector Command Failed
June 14, 2017 08:19
That bad part about these systems GE uses a power supply that is not adjustable and the slightest change in resistance on a connector the supply drops below 4.85 which locks up the system. I have replaced the supply with on which I can adjust up to 5.15 or you have to trace the supply wires and find out where it is dropping and repair it
to rate and post a comment
June 12, 2017 10:19
to rate and post a comment
CHRIS NALMPANTIS
re: GE Prodigy Detector Command Failed
May 15, 2014 05:47
check what detector board you have.some times the 45000 board with the altera is defect
to rate and post a comment
Desmond Johnson
re: GE Prodigy Detector Command Failed
May 15, 2014 05:47
Check the Error Log for System Reset Errors. If you have these move the arm to the center of the table, and the Source Tube to the center of the X axis and try again. It could be a cable. Also look at the molex on the low voltage supply. It gets flakey.
to rate and post a comment