GE Prodigy "Detector Command Failed"

Tags: Bone Densitometer, GE

(38)
(6)

John Mendoza

GE Prodigy "Detector Command Failed"

May 14, 2014 12:06

I have a GE Prodigy that gives me a Detector Command Failed error every time I do a QA. Any help or suggestions would be greatly appreciated.
Thanks
j.mendoza@scmes.com

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: GE Prodigy Detector Command Failed

May 15, 2014 08:04

measure the power supplies on the detector motherboard. many times the +5 vdc supply will be low

Log inor Register

to rate and post a comment

(3)

Tran Tuan

re: re: GE Prodigy Detector Command Failed

June 08, 2017 10:42

I check the voltage at TP38 and TP43 is 4.773VDC, Any ideas? thank you!

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: GE Prodigy Detector Command Failed

June 08, 2017 11:03

Reseat the connectors on the CSBC and Detector motherboard. There is also a connector in the wiring bundle where the upper and lower bundles are going to the tube and detector that causes issues

Log inor Register

to rate and post a comment

(3)

Tran Tuan

re: re: re: re: GE Prodigy Detector Command Failed

June 09, 2017 08:22

I reseat the connectors between csbc and DIB board already, but problem can not fix!

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: re: re: GE Prodigy Detector Command Failed

June 09, 2017 03:24

What is the power supply putting out?

Log inor Register

to rate and post a comment

(3)

Tran Tuan

re: re: re: re: re: re: GE Prodigy Detector Command Failed

June 13, 2017 09:34

I measure the power supply is 4.995VDC.

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: re: re: re: re: GE Prodigy Detector Command Failed

June 14, 2017 08:19

That bad part about these systems GE uses a power supply that is not adjustable and the slightest change in resistance on a connector the supply drops below 4.85 which locks up the system. I have replaced the supply with on which I can adjust up to 5.15 or you have to trace the supply wires and find out where it is dropping and repair it

Log inor Register

to rate and post a comment

This Comment has been deleted

June 12, 2017 10:19

Log inor Register

to rate and post a comment

CHRIS NALMPANTIS

re: GE Prodigy Detector Command Failed

May 15, 2014 05:47

check what detector board you have.some times the 45000 board with the altera is defect

Log inor Register

to rate and post a comment

Desmond Johnson

re: GE Prodigy Detector Command Failed

May 15, 2014 05:47

Check the Error Log for System Reset Errors. If you have these move the arm to the center of the table, and the Source Tube to the center of the X axis and try again. It could be a cable. Also look at the molex on the low voltage supply. It gets flakey.