The scanner reset errors could be caused by a number of issues, mostly low voltage related. To narrow down the culprit please send the full error log to us. info@dexasolutions.com
Measure the +5vdc supply on the DMB Detector Motherboard, these systems are know for connector issues and if it measures less than 4.85vdc the cpu chips will not function. Reseat the following connectors J4 on the CSBC, J6 on the DMB and trace the cable off the CSBC back to the where the upper bundle separates from the main harness from CSBC there is sometimes a connector in the cabel track
This is usually caused by the upper bundle cable . To be sure you can move the arm a few feet in any direction as well as unplug bulkhead 1 and 2 , usually the error will clear and computer will communicate with scanner. This error is usually intermittent and almost always is the upper bundle cable. Hope this helps
Assuming you've shut down the box and pc and powered them back on and still getting this. There is a company called Dexa Solutions whom offers free technical support and can help decipher logs and troubleshoot over the phone. (855)-844-3392
this is normally caused by the +5vdc on the DIB/DMB being too low (4.85 vdc or lower) it is normally due to the connectors on the CSBC and the ones going to the DIB/DMB. on the arm that supports the detector left side near the CSBC are several connectors, spray contact cleaner on them and reseat them several times
DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com
christy mcarthy
GE prodigy Scanner reset error
March 16, 2017 10:04
Hi every one ,
we have a scanner reset error on GE Prodigy dexa.
error log shows : 26533 DMB reset line active.
would appreciate any input
thanks
christy
to rate and post a comment
Ben Williams
re: GE prodigy Scanner reset error
March 17, 2017 08:17
HI Christy,
The scanner reset errors could be caused by a number of issues, mostly low voltage related. To narrow down the culprit please send the full error log to us. info@dexasolutions.com
Thanks,
Ben
to rate and post a comment
Lenny Place
re: GE prodigy Scanner reset error
March 17, 2017 08:17
Measure the +5vdc supply on the DMB Detector Motherboard, these systems are know for connector issues and if it measures less than 4.85vdc the cpu chips will not function. Reseat the following connectors J4 on the CSBC, J6 on the DMB and trace the cable off the CSBC back to the where the upper bundle separates from the main harness from CSBC there is sometimes a connector in the cabel track
to rate and post a comment
Matthew Larkins
re: GE prodigy Scanner reset error
March 17, 2017 08:17
This is usually caused by the upper bundle cable . To be sure you can move the arm a few feet in any direction as well as unplug bulkhead 1 and 2 , usually the error will clear and computer will communicate with scanner. This error is usually intermittent and almost always is the upper bundle cable. Hope this helps
to rate and post a comment
Cristiano Canova
re: GE prodigy Scanner reset error
March 17, 2017 08:17
Christy,
Mail me: cristiano.canova@yahoo.com.br
to rate and post a comment
Brad
re: GE prodigy Scanner reset error
March 17, 2017 01:23
Assuming you've shut down the box and pc and powered them back on and still getting this. There is a company called Dexa Solutions whom offers free technical support and can help decipher logs and troubleshoot over the phone. (855)-844-3392
to rate and post a comment
tim cowell
re: GE prodigy Scanner reset error
March 17, 2017 01:23
what is your email address. I can send you some information. tcowell@mmchs.org
to rate and post a comment
hakan bayram
re: re: GE prodigy Scanner reset error
August 19, 2022 11:18
O have same issue,if you could send mem the same information about the this issue, İ'll ne greatfullu TL you
to rate and post a comment
Lenny Place
re: re: re: GE prodigy Scanner reset error
August 22, 2022 10:55
this is normally caused by the +5vdc on the DIB/DMB being too low (4.85 vdc or lower) it is normally due to the connectors on the CSBC and the ones going to the DIB/DMB. on the arm that supports the detector left side near the CSBC are several connectors, spray contact cleaner on them and reseat them several times