GE ProSpeed 550 V Fail

March 23, 2006 08:26

Hello:
I have a couple of problems with a GE ProSpeed CT Scanner, one of them is having Led DS6 550 V Fault ON on Rotor Board. When performing Autotest S12 for Rotor Board is OK, however it doesn't go trough with S14 (when performing acc, run and brake on Rotor)Is there any way to check the 550 V? The PDU is model 9180. I will appreciatte any coments. thank you.
Best Regards,
E. Valdes

Log inor Register

to rate and post a comment

(57)
(3)
(6)

R.K.S singh

Re: GE ProSpeed 550 V Fail

March 26, 2006 05:40

Respected sir
i m also ge ct engg. sir u can check 550v supply
in power brush side
hope so u r geting xg error 201 or 103
can u tell me error code than i can help u this type of problem i had slove lot of time any dout u can call me my no +919392497167
thanks
rohit singh
india

Log inor Register

to rate and post a comment

(58)
(6)

Elias Valdes

Re: GE ProSpeed 550 V Fail

March 27, 2006 03:28

ERR 117 code 103 and 102

Log inor Register

to rate and post a comment

(4)

Mitesh Bhagat

Re: GE ProSpeed 550 V Fail

April 11, 2006 02:13

Hi,
Please check if your CTVRC module is ok ? Also check both IGBTs on left and right side below CTVRC.
Regards

Mitesh Bhagat
Mumbai,
India
miteshmails@rediffmail.com
+91 9323 22 1703

Log inor Register

to rate and post a comment

(243)
(18)
(9)

Carlos Duran

Re: GE ProSpeed 550 V Fail

April 05, 2006 02:40

hello:

the led in rotor board is normally ON if you not give a command to put on the PDU.

the error 117 "XG error", is given if your Xray generator looses the 550VDC, but keep in mind that is measured on CTVRC module.

if you have more questions, please contact me.

bye

Log inor Register

to rate and post a comment

(6)

m.k.karthickk vasu

Re: GE ProSpeed 550 V Fail

June 27, 2008 09:21

Hello

Is any error LED glows in common PDU PCB ? pls check..

Log inor Register

to rate and post a comment