Over 150 Missouri Auctions End Today - Bid Now

GE PROSPEED

(46)

(5)

Charles Psarreas

GE PROSPEED

June 11, 2004 08:26

THE SYSTEM IS TRANSFERING IMAGES TO AND FROM OPTCAL DISK OK, BUT WHEN I STORE TO OPTICAL DISK IT DOES NOT PUT THE S TO FLAG IT IS DONE.

WHEN I BRING IT BACK FROM THE OPTICAL DISK TO THE HARD DISK THE R FLAG IS ON.

ALL DATA IS GOOD , BUT S FLAG IS NOT THERE.

Log inor Register

to rate and post a comment

(47)

Mike Lynch

Re: GE PROSPEED

June 14, 2004 12:53

If it does it with only that Optical disk, it may be that it needs reformatting again. If it does it with all disks, then it is time for you reload software. There is a software fix for this but doing the LFC is safer and takes just a little more time.

Log inor Register

to rate and post a comment

(50)
(3)
(3)

Iftikhar Khan

Re: GE PROSPEED

June 23, 2004 12:58

we have a very experienced engineer currently in NY. he can help u in this regard , call 16462018203.