GE Proteus intermittent error 60 -802

Tags: Radiology - R/F, GE

(16)
(1)

Dick Lee

GE Proteus intermittent error 60 -802

November 28, 2018 08:49

Exposure backup time exceeded > but all exposures are manual.

Occurs about 1-2/week.

Repeat needed> following exposure>ok, no error.

Thank you.

Log inor Register

to rate and post a comment

(72)
(18)

tim cowell

re: GE Proteus intermittent error 60 -802

November 29, 2018 09:43

try reseating all connections on the generator I/F board.

Log inor Register

to rate and post a comment

(16)
(1)

Dick Lee

re: re: GE Proteus intermittent error 60 -802

November 30, 2018 09:36

Hi Tim,
thanks. I will check.
You have seen this error before?

Log inor Register

to rate and post a comment

(118)
(16)

VENGE

re: GE Proteus intermittent error 60 -802

November 29, 2018 09:43

From JEDI Error Codes:
Potential cause:
1. System
2. System-Generator cable
3. I/F board: exposure line stuck to the active state

Recommended action/ Troubleshooting guide:
- Retry, changing parameters and duration
- Disconnect system-Generator cable in different
places and check exposure command line voltage

Log inor Register

to rate and post a comment

(16)
(1)

Dick Lee

re: re: GE Proteus intermittent error 60 -802

November 30, 2018 09:37

Hi Venge,

thanks for the info...
have you seen this error before?

It's odd because AEC is not selected, and the error is about backup time exceeded.

Log inor Register

to rate and post a comment

DMEDS

re: GE Proteus intermittent error 60 -802

December 01, 2018 01:18

i have manuals
magmesme@gmail.com
regards

Log inor Register

to rate and post a comment

(2)

(1)

george blaser iv

re: GE Proteus intermittent error 60 -802

July 20, 2020 11:32

Were you able to fix this error? I'm having similar issue.

Log inor Register

to rate and post a comment

(10)

(1)

Rupert Theroue

re: GE Proteus intermittent error 60 -802

August 25, 2020 10:56

60-802 : Exposure backup time exceeded - the exposure command lasted longer than the duration that was loaded by the system (Backup time +5%)

Potential Causes:
1. Exposure command line stuck in an active state
2. PPC kV Control Board

Troubleshooting:
1. Reset and retry after changing parameters and duration.
2. If error persists, replace kV Control board.

You indicated in a previous answer this is occurring even when AEC is not selected. As such, I would highly recommend you order, replace, and reload all firmware to the kV Control board. You cold also try replacing the kV Control Board NVRAM and reloading all firmware.

Hope you have a good backup of the system available.