Standard Service manual for the system has Cal 009 for the touch screen, There should be a cable that came with the system to accomplish this procedure.
1. Prepare or find communication cable according to illustration.
2. Prepare a standard computer keyboard PS2.
3. Obtain system software disk, boot disk.
4. Open the back cover of the system console.
5. Use the communication cable connecting the keyboard and the console connector CN1.
6. Insert the System software, boot disk into console.
7. Reset the console or turn the power on.
8. Run “c:/Tbdriver.exe"
9. Run “Tbcal.exe”, then operate according the instruction.
10. main menu "H" for hard calibrate.
11. touch the extreme lower left corner and the right corner, then press any key to return to main menu.
12. enter "6" (640X480 graphics)
13. touch the center of the crosses twice then any key to get main menu
14 enter "7" test screen.
15 enter "x" to exit
16 cycle power
17. Mount the console cover after calibration.
Sell your scrap MRI magnets! Top prices, free shipping, and fast pickups. Don't let dust-collecting, idle assets go to waste. Contact: ryan.swiderski@cyclicmaterials.earth or call 437-826-1621
Kyle Schwartz
GE Proteus Touch Screen calibration
August 06, 2010 09:03
I need to locate the procedure for calibrating the proteus touch screen. Can anyone email it to me?
to rate and post a comment
Staate Hayward
Re: GE Proteus Touch Screen calibration
August 09, 2010 05:06
Standard Service manual for the system has Cal 009 for the touch screen, There should be a cable that came with the system to accomplish this procedure.
1. Prepare or find communication cable according to illustration.
2. Prepare a standard computer keyboard PS2.
3. Obtain system software disk, boot disk.
4. Open the back cover of the system console.
5. Use the communication cable connecting the keyboard and the console connector CN1.
6. Insert the System software, boot disk into console.
7. Reset the console or turn the power on.
8. Run “c:/Tbdriver.exe"
9. Run “Tbcal.exe”, then operate according the instruction.
10. main menu "H" for hard calibrate.
11. touch the extreme lower left corner and the right corner, then press any key to return to main menu.
12. enter "6" (640X480 graphics)
13. touch the center of the crosses twice then any key to get main menu
14 enter "7" test screen.
15 enter "x" to exit
16 cycle power
17. Mount the console cover after calibration.