Over 500 Total Lots Up For Auction at One Location - NJ 06/15

ge proteus x.ray

Tags: Radiology - R/F

Posted by an Anonymous User

ge proteus x.ray

July 10, 2015 03:49

I looking for service manual of x.ray machine ge Proteus RX/a or help for error 110,903

Log inor Register

to rate and post a comment

(12)

Michael Garnica

re: ge proteus x.ray

July 11, 2015 09:22

error code means Tube exceeded 70 degree C it got too hot. Thermal error code and 110 804 Tank thermal error

Log inor Register

to rate and post a comment

(265)
(50)
(28)

Wayne Horsman

re: ge proteus x.ray

July 11, 2015 09:22

what is your email address?

Log inor Register

to rate and post a comment

(10)

Mohammad irfan

re: re: ge proteus x.ray

July 12, 2015 02:20

Irfanrandhawa@yahoo.com

Log inor Register

to rate and post a comment

(265)
(50)
(28)

Wayne Horsman

re: ge proteus x.ray

July 11, 2015 09:22

110 => Thermal error, Class 5, Code 0903H, Tube exceeded 70degC

Log inor Register

to rate and post a comment

(9)

Engr Ali Hussain

re: ge proteus x.ray

July 11, 2015 09:27

Hi
GE have online free library of all manuals. You can download it online..

Just google GE Healthcare manual.

Thank you.

Log inor Register

to rate and post a comment

(51)
(15)

Kamran Butt

re: ge proteus x.ray

July 11, 2015 09:30

Download it from ge support documentation library.,

Log inor Register

to rate and post a comment

(12)
(8)

Arturo Papa

re: ge proteus x.ray

July 11, 2015 09:30

110-
0903H
70 degree C thermal
safety error
70° loop detected open
1. X-ray tube too hot;
normal error
2. Wiring problem
3. Sensor problem (Tube)
4. rotation board.

IT COULD BE A BROKEN WIRE OR DEFECT TEMP. SENSOR.
CHECK CONTINUITY (OHM)
CHECK ON THIS SITE THERE IS THE SERVICE MANUAL:
http://apps.gehealthcare.com/servlet/ClientServlet?REQ=RNEW&MODALITY=XR

2 people liked this
Log inor Register

to rate and post a comment

(10)

Mohammad irfan

re: ge proteus x.ray

July 12, 2015 02:20

Thanks to all thermal sensor and tube temp is normal and wiring OK some time error appeared

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Multi-Vendor Imaging Service Partner

BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729

stats

Upcoming Auction Events

Therapeutic Technologies
TIMED