System having Error 04 arm fatal failure and Error 55 tube housing angulation.
Also when I tried going back to tube heater calibration it gave me Out of Order.
How do we be able to access again or reset the tube tilt page?
On error log, DMR error codes are 05701 and 16110.
It sounds like you lost communication or calibration from the generator. I would check all serial cables and connectors going to the generator after I shut down the generator first. I would re-load my cal files from the floppy second. Then I will Proceed to the solutions below if reloading cal files didn't help.
I personally saw one of the generator connectors was a few millimeters out, pushed it in and tightened screws and everything start to work again.
For Error 04: If the problem occurs during boot,check optical fiber between Gantry interface board and
generator, also can bus. If the problem occurs when requesting an exposure, check the Bucky.
For the tube housing angulation Error: Try to re-calibrate using Job Card IST 029.
Did checked all cable and connectors all seems ok. Unfortunately no floppy diskettes on site or any backup, customer just bought a new Desktop so no backup files on it.
Still getting error 04 after checking the bucky.
Did also the tube angulation calib,tube will angulate back and forth but still getting error 55.
I tried clicking on Tube Tilt but nothing happens on screen, it did not open up a window for tube tilt.
Any help is appreciated.
I would suggest to check with Debi of DigitecMedical, her email is debi.breland@digitecmedical.com, for sure they can definitely help you out if you have issues with GE Senographe. This company was able to revive our mammo unit and they are available 24x7.
If you need technical support please don't hesitate to contact us. My email is debi.breland@digitecmedical.com and you are welcome to email me and we will help you with this.
I need to obtain a new Software Key for a Seno 2000D that apparently has an expired Software Protection Key. Following the log in, and as the application software begins to load - the following message is displayed: Invalid software key for AWS Station. We checked and verified with the command: install.key that the default key is the same key that is written into the installation worksheet. This same security key code also appears on a label attached to the Sun Blade 150 computer. The computer was just replaced to resolve a previous issue on boot up, which stopped the auto-boot process at the very beginning, and displayed the message: IDPROM contents are invalid. The replacement Sun Blade 150 computer resolved this issue, allowing the computer to boot to the log in screen (the original unit's two hard drives were installed into the identical replacement Sun computer). The unit accepts the log in info and then starts loading the application screen, only for it to be interrupted with the message about the invalid security key. Can you folks sell us a replacement Software Key for this system? I can be reached on my Cell Phone at: (914) 850-9922 - Bill DeFalco
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
GE SENOGRAPHE 2000D
May 30, 2012 05:22
System having Error 04 arm fatal failure and Error 55 tube housing angulation.
Also when I tried going back to tube heater calibration it gave me Out of Order.
How do we be able to access again or reset the tube tilt page?
On error log, DMR error codes are 05701 and 16110.
to rate and post a comment
Muneer Abott
re: GE SENOGRAPHE 2000D
May 30, 2012 11:32
It sounds like you lost communication or calibration from the generator. I would check all serial cables and connectors going to the generator after I shut down the generator first. I would re-load my cal files from the floppy second. Then I will Proceed to the solutions below if reloading cal files didn't help.
I personally saw one of the generator connectors was a few millimeters out, pushed it in and tightened screws and everything start to work again.
For Error 04: If the problem occurs during boot,check optical fiber between Gantry interface board and
generator, also can bus. If the problem occurs when requesting an exposure, check the Bucky.
For the tube housing angulation Error: Try to re-calibrate using Job Card IST 029.
Good Luck!
to rate and post a comment
Gee Gicale
re: re: GE SENOGRAPHE 2000D
May 31, 2012 11:37
Did checked all cable and connectors all seems ok. Unfortunately no floppy diskettes on site or any backup, customer just bought a new Desktop so no backup files on it.
Still getting error 04 after checking the bucky.
Did also the tube angulation calib,tube will angulate back and forth but still getting error 55.
I tried clicking on Tube Tilt but nothing happens on screen, it did not open up a window for tube tilt.
Any help is appreciated.
to rate and post a comment
Gene Goryukhin
re: re: GE SENOGRAPHE 2000D
June 27, 2012 04:29
Hi. Does anyone have a service manual for 200D? Thanks
to rate and post a comment
Gee Gicale
re: re: re: GE SENOGRAPHE 2000D
June 27, 2012 05:26
I would suggest to check with Debi of DigitecMedical, her email is debi.breland@digitecmedical.com, for sure they can definitely help you out if you have issues with GE Senographe. This company was able to revive our mammo unit and they are available 24x7.
to rate and post a comment
Deborah Breland
re: GE SENOGRAPHE 2000D
May 31, 2012 10:03
If you need technical support please don't hesitate to contact us. My email is debi.breland@digitecmedical.com and you are welcome to email me and we will help you with this.
to rate and post a comment
Bill DeFalco
re: re: GE SENOGRAPHE 2000D
December 31, 2015 10:10
I need to obtain a new Software Key for a Seno 2000D that apparently has an expired Software Protection Key. Following the log in, and as the application software begins to load - the following message is displayed: Invalid software key for AWS Station. We checked and verified with the command: install.key that the default key is the same key that is written into the installation worksheet. This same security key code also appears on a label attached to the Sun Blade 150 computer. The computer was just replaced to resolve a previous issue on boot up, which stopped the auto-boot process at the very beginning, and displayed the message: IDPROM contents are invalid. The replacement Sun Blade 150 computer resolved this issue, allowing the computer to boot to the log in screen (the original unit's two hard drives were installed into the identical replacement Sun computer). The unit accepts the log in info and then starts loading the application screen, only for it to be interrupted with the message about the invalid security key. Can you folks sell us a replacement Software Key for this system? I can be reached on my Cell Phone at: (914) 850-9922 - Bill DeFalco
to rate and post a comment
jaime alberto
re: GE SENOGRAPHE 2000D
February 23, 2016 12:45
Good morning friend , I need the service manual please help me .