GE Senographe DMR -generator failure problem

Tags: Mammo, GE

Alexander Tecson

GE Senographe DMR -generator failure problem

January 23, 2013 09:34

Good day!

I have a problem with a GE Senographe DMR,

After boot-up the error displayed
at console "E01 GENERATOR FAILURE 007/228".

Any help and suggestion? thank you in advance,

1 person liked this
Log inor Register

to rate and post a comment

(64)
(14)
(82)

Jamal Bazzi

re: GE Senographe DMR -generator failure problem

January 24, 2013 09:11

replace the hv board located on the hv tank

Log inor Register

to rate and post a comment

(17)
(4)

Aamir Misbah

re: re: GE Senographe DMR -generator failure problem

September 10, 2013 10:33

i have problem with GE Senographe DMR.

after switching on the unit an error (150 lock jammed ) comes and hang the unit .can you help

Log inor Register

to rate and post a comment

(8)

worldisflat

re: re: GE Senographe DMR -generator failure problem

April 21, 2020 10:20

i have problem with e Senographe S61 error.
i turn on/off after completed boot aws system but persist error.
can you help me?

Log inor Register

to rate and post a comment

(24)
(9)

cristian oliveira

re: GE Senographe DMR -generator failure problem

January 24, 2013 09:35

good day alexander.

for this you need to access the device whith laptop and see which entry this failed.
If you don“t have the SW acess email my cristian.oliveiramedx@gmail.com

1 person liked this
Log inor Register

to rate and post a comment

Alexander Tecson

re: re: GE Senographe DMR -generator failure problem

February 06, 2013 01:16

Thank you for the reply, i already check the error log display,.

E01 Generator failure 007/228
Securities:
POWER_SUPPLY_FAIL XJ2 26: NO
EXPOSURE_TIME_MAX XJ2 27: NO
REGULATION_TIMEOUT XJ2 28 : YES
HEAT_SINX_TEMP XJ2 29 : NO
INVERTER_IMAX XJ2 30 : NO
KV_DROP XJ2 31: YES
KV_MAX XJ2 32: YES
MA_MAX XJ2 33 : YES

Does the line XJ2 28 / XJ2 31 / XJ2 32 / XJ2 33 is normal?
I already checked all the dc inputs, cables and fuses -all stable and good in condition.

I observed the DS 12 (SEC R) AND DS 13 (SEC BL) is power on, is it normal?

Thank you in advance.

Log inor Register

to rate and post a comment

(17)
(4)

Aamir Misbah

re: re: GE Senographe DMR -generator failure problem

September 10, 2013 10:33

i have problem with GE Senographe DMR.

after switching on the unit an error (150 lock jammed ) comes and hang the unit .can you help

Log inor Register

to rate and post a comment

(30)
(5)

ARTURO CORONA

re: re: GE Senographe DMR -generator failure problem

October 01, 2018 09:09

Could you help me I hace dmr whith error S2 exposing aborted by hv arcing. I don't have software for the dmr. My email
acoronap@hotmail.com

1 person liked this
Log inor Register

to rate and post a comment

(244)
(19)

Staate Hayward

re: GE Senographe DMR -generator failure problem

January 24, 2013 09:37

Gene init tests: one fatal security input, the 228 identifies the specific input.Access generator error log display for full error decoding. Do you have a laptop that you can use to monitor and emulate the system, ie. Mammocom software?

Log inor Register

to rate and post a comment

Alexander Tecson

re: re: GE Senographe DMR -generator failure problem

February 06, 2013 09:27

Thank you for the reply, i already check the error log display,.

E01 Generator failure 007/228
Securities:
POWER_SUPPLY_FAIL XJ2 26: NO
EXPOSURE_TIME_MAX XJ2 27: NO
REGULATION_TIMEOUT XJ2 28 : YES
HEAT_SINX_TEMP XJ2 29 : NO
INVERTER_IMAX XJ2 30 : NO
KV_DROP XJ2 31: YES
KV_MAX XJ2 32: YES
MA_MAX XJ2 33 : YES

Does the line XJ2 28 / XJ2 31 / XJ2 32 / XJ2 33 is normal?
I already checked all the dc inputs, cables and fuses -all stable and good in condition.

I observed the DS 12 (SEC R) AND DS 13 (SEC BL) is power on, is it normal?

Thank you in advance.

1 person liked this
Log inor Register

to rate and post a comment

(244)
(19)

Staate Hayward

re: re: re: GE Senographe DMR -generator failure problem

February 06, 2013 01:16

In my book all the Securities should be "NO" in the standby status. What were the "Generator Signals" listing, should be Y, Y, N, N, N, N, N, N, Y, N, N, N. In emulation mode HV test, Setup, Gene, Maint, Maint, Maint, HV, HV. In emulation mode mA test, Setup, Gene, Maint, Maint, Maint, Heat. staate.hayward@sodexo.com send results. Sounds like the High Voltage tank and 701 board are shot. Could also be 400 control board.

Log inor Register

to rate and post a comment

(28)
(15)

MIGUEL GONZALEZ

re: GE Senographe DMR -generator failure problem

January 25, 2013 10:40

you need the mammocom software and the advanced service manual

Log inor Register

to rate and post a comment

(54)
(7)
(9)

Deborah Breland

re: GE Senographe DMR -generator failure problem

February 06, 2013 01:16

Alexander. I believe that our engineers have a solution to your problem Please email me at debi.breland@digitecmedical.com and I will be happy to help you with this.

Log inor Register

to rate and post a comment

Moutaz Mesh

re: re: GE Senographe DMR -generator failure problem

April 05, 2015 10:35

I have an Error message E 172/056 and I believe that I have to recalib the compression paddle.
could you please help me with the service password..
thanks a lot

Log inor Register

to rate and post a comment

(244)
(19)

Staate Hayward

re: re: re: GE Senographe DMR -generator failure problem

April 06, 2015 10:28

These units have a double switch activation, One being the request to move and the second being "Yes there is an operator present". Both have to be active for operation of the paddle compression. Check your wiring, switches, cables and connections first.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Multi-Vendor Imaging Service Partner

BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729

stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)