GE SENOGRAPHE DMR HELP

Tags: Mammo, GE

(220)
(35)
(1)

Tatiana Bautista

GE SENOGRAPHE DMR HELP

February 03, 2021 10:18

Regards, after having replaced the battery of the generator we have removed the error of checksum installing the gene, after this procedure the following messages appear:
S16 Selected Film/screen is uncalibrated
S09 No compression detected
S15 Arm not present
was successful communicating with mammocom software with win98 PC and serial port with db9 cable, they can help me with calibrations. dat please the mammograph in DMR no DMR+, or can you help me calibrate the mammograph please, we should worry about the message Arm not present?. I will be immensely grateful, thank you in advance

Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

re: GE SENOGRAPHE DMR HELP

February 04, 2021 10:18

The arm is installed on the console, with the menu. Do you know how? That would take care of the compression. Then you have to calibrate the film/screens

Log inor Register

to rate and post a comment

(243)
(18)
(9)

Carlos Duran

re: GE SENOGRAPHE DMR HELP

February 10, 2021 10:20

When you remove the battery, all calibrations are lost.

You need to redo all of them; the easy way is via service PC

Log inor Register

to rate and post a comment