GE SENOGRAPHE DMR UPLOAD BACKUP

(220)
(35)
(1)

Tatiana Bautista

GE SENOGRAPHE DMR UPLOAD BACKUP

January 13, 2021 11:22

Regards, Can you help me with the correct procedure to upload the backup to the mammograph, should I put the mammograph in service mode? thanks in advance,

Log inor Register

to rate and post a comment

(105)
(17)
(17)

Doscar Ernesto Perez Pereyra

re: GE SENOGRAPHE DMR UPLOAD BACKUP

January 15, 2021 10:47

When you reload software backup it’s the most important time (Mostly Generator Backup). But you don’t change generator I just assume only lost you hard disks. In that case just upload your detector files and re calibrate it. About the Motions calibrations like Tilt and Height are very easy to do it.

Log inor Register

to rate and post a comment

(18)

Annas

re: GE SENOGRAPHE DMR UPLOAD BACKUP

January 15, 2021 10:47

Select service desktop >Utilities >restore >select component i.e AWS or IDC or DMR, one at one time, select backup either from AWS hard disk or from floppy or cd rom and press start. After all restore restart whole system. If you restore DMR you may see checksum error on xray console. Perform checksum from xray console using 5332 as . Procedure is given in service manual if you need mail me or check GE common document library.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Auction Alert - Live Now: Therapeutic Technologies Surplus Auctions June 15th

Therapeutic Technologies closeout auctions take place on DOTmed.com. Bid today on 538 lots of medical equipment. Inspections may be done by appointment. Hours are Monday through Friday 9am-4pm Contact chrisnuytens1@gmail.com to schedule.

stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)