tahar nouri

GE silouhette VR error message 60

August 09, 2011 05:43

hi to all

i have a a problem with silouhette VR , displays an

error message error 60 in the KV window .


thanks in advance

Log inor Register

to rate and post a comment

(259)
(59)
(3)

Jose Vallejos

Re: GE silouhette VR error message 60

August 10, 2011 10:12

Hi,

Check all component about generation of KV, (KV control, Inverter, Resonancy assy, tank HV, cable HV, tube Rx).

You can test with other techniques lower kv and mA and time.

Best regards
Jose

Log inor Register

to rate and post a comment

(158)
(30)
(1)

tahar nouri

Re: GE silouhette VR error message 60

August 10, 2011 05:33

HI ,

thanks for your replay; in order to check that

I need a service software for jedi geneartor ,

i would be very thankful ,if you could send me

a copy of service software.