G.E system five

Tags: Ultrasound, GE

Anil .

G.E system five

June 14, 2016 10:12

Dear All,

I have one G.E system five ultrasound machine.two issues. Monitor display intermittantly goes off. There is plenty of cross lines across scan image area.Some of the system tests are getting failed like Import board test,Long RFT test,Image port board test. Can anybody please suggest what could be the reasons and how we can sort out these issues.

My email Id is- saahithy@gmail.com
If anybody having the service mannual and trouble shooting guide for this particular can you please provide me.

Thanks,
Anil

Log inor Register

to rate and post a comment

(21)
(2)

Pablo Delgado

re: G.E system five

June 14, 2016 05:06

Hi,

What system are you talking about? Logiq 5 or logiqP5?
For Logiq P5, I think that the problem is related with SYSCON board where the video is generated to display in LCD, then you should check the CL1TRX assy where the RF signals are generated. This boards are labeled.
For Logiq 5, you have to check in BEP the VIC card where the video is generated.

Log inor Register

to rate and post a comment

Anil .

re: re: G.E system five

June 16, 2016 08:29

Dear Sir,

Thanks for your reply.It is Vingmed sound, system Five(Vivid 5)

Log inor Register

to rate and post a comment

(195)
(47)
(4)

EROTOKRITOS SKENTOS

re: G.E system five

June 15, 2016 11:48

Anil first of all tell me how you have done the tests since the monitor goes off.Secondly tell me if you can connect a pc through service port and read the messages there.In the system five there is a cover on the back and inside there is a parallel port for service.It is very important to connect through hyper terminal.

Log inor Register

to rate and post a comment

Anil .

re: re: G.E system five

June 16, 2016 08:29

Dear Sir,

Thanks for your reply.The problem with the monitor, frankly speaking i couldn't vitness it even after long testing. But the user is saying some time it happens.

There is an option for test in the machine inside system setup,which can be accessed through 'setup' switch available in the keypad.

Regarding connecting the pc through parellel port , i didnt done it. Can u please breif me how it can be done.

Thanks,
Anil

Log inor Register

to rate and post a comment

(195)
(47)
(4)

EROTOKRITOS SKENTOS

re: re: re: G.E system five

June 17, 2016 09:54

I will send you the procedure for connecting a pc with the system five but in order to see the messages you must be there when the problem occures.

Log inor Register

to rate and post a comment

Anil

re: re: re: re: G.E system five

June 19, 2016 03:01

Hi

Thanks for the reply .Waiting to here from you soon

Log inor Register

to rate and post a comment

stats
DOTmed text ad

BK and other Urological Systems available. BK Probe repair available!!!

Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net

DOTmed Certified stats
advertisement

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)