Dear All,
I have a problem in GE sytec sri CT , Anode over KV appears(ER-117 Code-537 followed by ER-101 Code-54), sometime on 80KV it exposes but on 120KV & 135 KV it resume. I have replaced the CT TUBE (Brand New), HV Tank, inverter, KV Board, OGP Board. But still the error is coming this is intermediate error. So can u suggest some technical help in this matter?
DEAR HAI
THIS IS ROHIT SINGH I M GE CT ENGG I HAD FACE THIS TYPE OF PROBLEM SO MAY TIMES IN SYTECH SRI AND SYTECH SNERGY SYSTEM BOTH R SAME SYSTEM
ERROR 117 537
means ANODE OVER KV
POSIBLITY
1 X-RAY TUBE SPITS PERFORM HEAT SOAK PROCEDURE
2.KV BOARD IN E RACK
3. OPTICAL CABLE FROM E RACK TO HV INVERTER
4. HV TANK ON INVERTER SECTION
SOME TIMES THE PTOBLEM MAY BE OF SLIP RING WHERE 550 V POWER BRUSH R IN LOSSE CONT CHANGE THEM
HAD U CHANGE PARMETER IN CPU BOARD WHILE CHANGING TUBE AND HV TANK U R SURE THAT THE TANK ASSY U R USING UR INVERTER IS GOOD .
MORE I CAN TELL WHEN U GIVE ME BREIF MAIL PLEASE EXPLAINE EVERY THING WHAT U HAD DONE
IF ANY TYPE OF PROBLEM IN PROSPEED AND HISPEED PLEASE FEEL FREE TO MAIL ME OR I CAN COME AND SLOVE UR PROBLEM SPARE ALSO AVALIBLE
Dear
your ER 101 (54)
X-RAY RESPONCE TIME OUT IT IS DUE TO WHEN U HAD GIVE THE COMMAND AND YOUR GENATOR IS NOT REAY AND YOUR OGP HAD GIVEN FEED BACK TO CPU NO WE CANNOT SO THIS APEEARS
BUT THE COUASE IS AS FOLLOS
1 HOT GANTRY
2.PDU XFMR CONNECTOR
3 F BOARD IN E RACK
HAD U CHANGE CPU BOARD IN OBC THAT IS SUSPECTED
Online auction of modern high-end laboratory and biopharma research equipment from the recent bankruptcy of Spexis AG in Allschwil near Basel, Switzerland. Start bidding today! closes on June 4. info@dome-auctions.com..+31(0)20 808 2880
asrar mustafa
Ge Sytec Sri CT Scan problem
March 13, 2008 11:29
Dear All,
I have a problem in GE sytec sri CT , Anode over KV appears(ER-117 Code-537 followed by ER-101 Code-54), sometime on 80KV it exposes but on 120KV & 135 KV it resume. I have replaced the CT TUBE (Brand New), HV Tank, inverter, KV Board, OGP Board. But still the error is coming this is intermediate error. So can u suggest some technical help in this matter?
Regards
Asrar Mustafa
to rate and post a comment
R.K.S singh
Re: Ge Sytec Sri CT Scan problem
March 13, 2008 02:19
DEAR HAI
THIS IS ROHIT SINGH I M GE CT ENGG I HAD FACE THIS TYPE OF PROBLEM SO MAY TIMES IN SYTECH SRI AND SYTECH SNERGY SYSTEM BOTH R SAME SYSTEM
ERROR 117 537
means ANODE OVER KV
POSIBLITY
1 X-RAY TUBE SPITS PERFORM HEAT SOAK PROCEDURE
2.KV BOARD IN E RACK
3. OPTICAL CABLE FROM E RACK TO HV INVERTER
4. HV TANK ON INVERTER SECTION
SOME TIMES THE PTOBLEM MAY BE OF SLIP RING WHERE 550 V POWER BRUSH R IN LOSSE CONT CHANGE THEM
HAD U CHANGE PARMETER IN CPU BOARD WHILE CHANGING TUBE AND HV TANK U R SURE THAT THE TANK ASSY U R USING UR INVERTER IS GOOD .
MORE I CAN TELL WHEN U GIVE ME BREIF MAIL PLEASE EXPLAINE EVERY THING WHAT U HAD DONE
IF ANY TYPE OF PROBLEM IN PROSPEED AND HISPEED PLEASE FEEL FREE TO MAIL ME OR I CAN COME AND SLOVE UR PROBLEM SPARE ALSO AVALIBLE
THANKING YOU
ROHIT SINGH
rohitsingh4u@hotmail.com
to rate and post a comment
R.K.S singh
Re: Ge Sytec Sri CT Scan problem
March 13, 2008 02:28
Dear
your ER 101 (54)
X-RAY RESPONCE TIME OUT IT IS DUE TO WHEN U HAD GIVE THE COMMAND AND YOUR GENATOR IS NOT REAY AND YOUR OGP HAD GIVEN FEED BACK TO CPU NO WE CANNOT SO THIS APEEARS
BUT THE COUASE IS AS FOLLOS
1 HOT GANTRY
2.PDU XFMR CONNECTOR
3 F BOARD IN E RACK
HAD U CHANGE CPU BOARD IN OBC THAT IS SUSPECTED