Over 500 Total Lots Up For Auction at One Location - NJ 06/15

GE SYTEC SYNERGY problem

(5)

Gantulga Chuluun

GE SYTEC SYNERGY problem

October 04, 2010 02:27

Dear Friends,

Please help me to figure out a problem with CT Scanner ,
error #28034

XG ERROR: SYSTEM SAFETY LOOP OPEN
XG STATE: STANDBY

slip ring,presure sensor,temperature sensor,all 3 sensors and connections checked ,
on the rotor board : tube1 led on,550v fault

Thanks,
Gantulga

Log inor Register

to rate and post a comment

(373)
(66)
(15)

yuan bo

Re: GE SYTEC SYNERGY problem

October 04, 2010 11:20

i think you should check the POWER SUPPLY OF 550v, it's easy to resolve.

Log inor Register

to rate and post a comment

(5)

Gantulga Chuluun

Re: GE SYTEC SYNERGY problem

October 05, 2010 10:07

the power supply 550v is ok
550v contactor not activating

Log inor Register

to rate and post a comment

(75)
(8)
(7)

JIFFY PAUL AKKARAKARAN

Re: GE SYTEC SYNERGY problem

October 05, 2010 02:19

Hi,
Please check the gantry right side cover switches located on the lower right side of the gantry cover then slide the service plate and tighten the wing nut. If is is ok check whether the CTVRC module (anode rotation is working), if this is ok then check the filament board error leds. then check the HV inverter module. The 550v contactor will switch on only if there are no errors in the XG section and the console and the cover switches are closed. The system saftey circiut is connected with a series of switches starts from PDU goes to console then it goes through the right gantry cover switch, then passes through the slip rings then if XG is ready it returns again through the slip rings and reaches the PDU. So check this loop.

Good luck

Jiffy Paul Akkarakaran

Log inor Register

to rate and post a comment

stats
DOTmed text ad

MRI & CT PARTS USED-TESTED, REFURBISHED. Call (512) 477-1500

Extensive inventory of Philips & Siemens MRI & CT used-tested, refurbished parts. Knowledgeable Sales and Inventory Team. 90-Day Warranty. Bilingual Support. Competitive Pricing. Technical Support. ISO Certified. parts@keimedicalimaging.com

DOTmed Certified stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)