Mohamad Ismail
GE Sytec Synergy problem
May 06, 2005 05:49
Dear All , i have a problem in GE Synergy CT , Anode Over KV appear every time i press start button, sometime on 80KV it expose but on 120KV it resume. I have replaced the inverter,FUSES without change.
Can anybody help ?
to rate and post a comment
shahzad Ahmad
Re: GE Sytec Synergy problem
May 06, 2005 04:19
Check the KV calibration.
to rate and post a comment
Kakhaber Gvaramia
Re: GE Sytec Synergy problem
May 07, 2005 04:51
Dear Sir,
At present I have the same problem . It is not Inverter's deffect. Check the HV Tank. ask somebody to do the stationary scan on 120KV and check if it is hearing noice from HV tank . If yes , you should replace HV tank . Otherwise , you have problem with the tube.
K.Gvaramia
to rate and post a comment
SANJIV GOYAL
Re: GE Sytec Synergy problem
June 13, 2005 10:53
We had similiar problem. It could be solved by changing the HV tank oil. You will need to clean the inside throughly of all the carbon deposits due to arching. That should solve your problem.