Over 500 Total Lots Up For Auction at One Location - NJ 06/15

GE Sytec Synergy problem

(29)
(5)

Mohamad Ismail

GE Sytec Synergy problem

May 06, 2005 05:49

Dear All , i have a problem in GE Synergy CT , Anode Over KV appear every time i press start button, sometime on 80KV it expose but on 120KV it resume. I have replaced the inverter,FUSES without change.
Can anybody help ?

Log inor Register

to rate and post a comment

(39)

(3)

shahzad Ahmad

Re: GE Sytec Synergy problem

May 06, 2005 04:19

Check the KV calibration.

Log inor Register

to rate and post a comment

(49)

Kakhaber Gvaramia

Re: GE Sytec Synergy problem

May 07, 2005 04:51

Dear Sir,

At present I have the same problem . It is not Inverter's deffect. Check the HV Tank. ask somebody to do the stationary scan on 120KV and check if it is hearing noice from HV tank . If yes , you should replace HV tank . Otherwise , you have problem with the tube.


K.Gvaramia

Log inor Register

to rate and post a comment

SANJIV GOYAL

Re: GE Sytec Synergy problem

June 13, 2005 10:53

We had similiar problem. It could be solved by changing the HV tank oil. You will need to clean the inside throughly of all the carbon deposits due to arching. That should solve your problem.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Source for Sony Medical Displays, Printers & More!

Ampronix, a Top Master Distributor for Sony Medical, provides Sales, Service & Exchanges for Sony Surgical Displays, Printers, & More. Rely on Us for Expert Support Tailored to Your Needs. Email info@ampronix.com or Call 949-273-8000 for Premier Pricing.

stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)