Hello all. I have a problem with transducer M5SC-D on GE S70 ultrasound. Image quailty is not so good. I contacted two suppliers, I have sent them image example and they both said that the problem is in weak crystals. It's strange because transduces is only 4 years old. They both said that repair is too expensive and the new one is bette option. Can someone check image from the link to see is it really weak crystals problem? The same picture was sent to them. Thank you in advance.
there are maybe tree or four problems in any transducer
air bubble below surface ; dirty at the connection; crystals poor or broken; mesh of hearth (copper) defective .
FIRST CLEAN THE CONNECTOR WITH ALCOHOL ISOPROPILIC 99% PURE REACTIVE DEGREE (SPRAY)
Sell your scrap MRI magnets! Top prices, free shipping, and fast pickups. Don't let dust-collecting, idle assets go to waste. Contact: ryan.swiderski@cyclicmaterials.earth or call 437-826-1621
Nevus Solaris
GE transducer M5SC-D problem
March 07, 2024 01:29
Hello all. I have a problem with transducer M5SC-D on GE S70 ultrasound. Image quailty is not so good. I contacted two suppliers, I have sent them image example and they both said that the problem is in weak crystals. It's strange because transduces is only 4 years old. They both said that repair is too expensive and the new one is bette option. Can someone check image from the link to see is it really weak crystals problem? The same picture was sent to them. Thank you in advance.
to rate and post a comment
Roberto Rojas
re: GE transducer M5SC-D problem
March 08, 2024 10:12
there are maybe tree or four problems in any transducer
air bubble below surface ; dirty at the connection; crystals poor or broken; mesh of hearth (copper) defective .
FIRST CLEAN THE CONNECTOR WITH ALCOHOL ISOPROPILIC 99% PURE REACTIVE DEGREE (SPRAY)
SECOND INSPECT THE SURFACE AND THE CABLE
THREE CHANGE THE SURFACE
FOUR BUY ANOTHER TRANSDUCER