GE transducer M5SC-D problem

Tags: Cardiac Imaging

(1)

Nevus Solaris

GE transducer M5SC-D problem

March 07, 2024 01:29

Hello all. I have a problem with transducer M5SC-D on GE S70 ultrasound. Image quailty is not so good. I contacted two suppliers, I have sent them image example and they both said that the problem is in weak crystals. It's strange because transduces is only 4 years old. They both said that repair is too expensive and the new one is bette option. Can someone check image from the link to see is it really weak crystals problem? The same picture was sent to them. Thank you in advance.

Log inor Register

to rate and post a comment

(37)
(6)

Roberto Rojas

re: GE transducer M5SC-D problem

March 08, 2024 10:12

there are maybe tree or four problems in any transducer
air bubble below surface ; dirty at the connection; crystals poor or broken; mesh of hearth (copper) defective .

FIRST CLEAN THE CONNECTOR WITH ALCOHOL ISOPROPILIC 99% PURE REACTIVE DEGREE (SPRAY)

SECOND INSPECT THE SURFACE AND THE CABLE

THREE CHANGE THE SURFACE

FOUR BUY ANOTHER TRANSDUCER

1 person liked this