Juan de la Cruz Martinez

GE VCT 64

May 25, 2014 11:28

I have a problem with this VCT doing queries from the work station, Dicom ping and regular ping works but when you try to query the VCT you don’t get any response. A load from cold solves this problem for about 3 to 4 days, but then the problem comes back. I have replace HD and memories on this VCT, all other functions work including pushing the images to the work station. Any ideas? This problem is going to drive me crazy. Thanks.

Log inor Register

to rate and post a comment

Charlie Britt

re: GE VCT 64

May 26, 2014 12:39

Have you tried a direct cable connection from the workstation to the VCT? . May be another IP or port in the LAN has the same IP ?
Try it...

Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

re: re: GE VCT 64

May 26, 2014 07:44

yes I did, I try cable connection to the Work station with the switch and then with direct croosover cable computer to computer and is the same.