I have this equipment that has problems only with open coils but in the close ones (like the coil for brain or knee)the equipment works fine.
The image is realy bad for open coils, very noisy and granular, but is excelent for the close coils.
Any idea will help.
I still have this problem. In fact, mydoubt is that in a study without coil (only using the body coil) the equipment works fine, but i have the problem in the open coils (like lumbar coil).
Please perform noise check, if you have a noise source, receive coils are more sensitive. Also granulation on Lumbar coils should be releated to Eddy Currents, please check it. At the last preamplifier should be suspect if the above items are ok.
Hi Tayfun, thanks for your answer. I am using the coil named Standard 13x30cm for lumbar. I perform the noise check and nothing come out. This coil is the only one that doesn´t work, but i probe this coil in another equipment and this works fine.
I dont know what else can do.
coil 13x30cm = don´t works (but works in other equipment)
Mario Silva
GE VECTRA - problem with coils
December 19, 2008 03:50
I have this equipment that has problems only with open coils but in the close ones (like the coil for brain or knee)the equipment works fine.
The image is realy bad for open coils, very noisy and granular, but is excelent for the close coils.
Any idea will help.
Thanks
to rate and post a comment
Mario Silva
Re: GE VECTRA - problem with coils
December 20, 2008 10:29
I still have this problem. In fact, mydoubt is that in a study without coil (only using the body coil) the equipment works fine, but i have the problem in the open coils (like lumbar coil).
Thanks
to rate and post a comment
Tayfun Canli
Re: GE VECTRA - problem with coils
December 20, 2008 04:35
Please perform noise check, if you have a noise source, receive coils are more sensitive. Also granulation on Lumbar coils should be releated to Eddy Currents, please check it. At the last preamplifier should be suspect if the above items are ok.
to rate and post a comment
Mario Silva
Re: GE VECTRA - problem with coils
December 21, 2008 07:48
Hi Tayfun, thanks for your answer. I am using the coil named Standard 13x30cm for lumbar. I perform the noise check and nothing come out. This coil is the only one that doesn´t work, but i probe this coil in another equipment and this works fine.
I dont know what else can do.
coil 13x30cm = don´t works (but works in other equipment)
all other coils = work fine
thanks