May 17, 2019 09:00

Regards, please help, we have a Ge Vivid 3 ultrasound that has the damaged software, we have access to its files, there is some way to rescue the licenses and know what software version it has, when starting the computer it gives an error cant load mux board, please help, thanks in advance, sincerely ing.lauragc@hotmail.com

Log inor Register

to rate and post a comment

(165)
(67)
(2)

Sergey Zheleznyak

re: GE VIVID 3 DAMAGED SOFTWARE MUX ERROR

May 20, 2019 09:27

Hello, Laura
Licenses are stored in LPT HASP dongle - can't be lost in most cases
Software version can be see via path C:\Dolphin\init\Dolphin.ini

The wrong type MUX might be set in
D:\Dolphin\Init\DolphinHWCfg.ini

Best Regards