Aram Chilingaryan
GE Vivid 7 does not start up
January 07, 2018 09:51
Hi All,
Ge Vivid 7 does not start up. When I switch it on the yellow light on the power button appears for a few seconds and then goes off.
Equipment has not used for one year. Before that it was ok. What could be a problem?
Thanks
to rate and post a comment
tim cowell
re: GE Vivid 7 does not start up
January 08, 2018 09:35
Check BEP power supply
to rate and post a comment
Aram Chilingaryan
re: re: GE Vivid 7 does not start up
January 08, 2018 01:54
Thank you for your reply Tim.
One specialist checked BEP and said that it is ok. Is there any other possible causes?
to rate and post a comment
Manjul Kumar
re: re: re: GE Vivid 7 does not start up
January 11, 2018 08:55
Check BEP a.c. input voltage and also check a.c. d c distributor box
to rate and post a comment
tim cowell
re: re: re: GE Vivid 7 does not start up
January 11, 2018 01:07
There is a purple wire at the BEP that supplies 5Vdc to the power switch for the amber light. If you do not have 5Vdc at this wire then your power supply is bad.