joao albuquerque
GE Vivid I - ECG Trace Boot Error
December 07, 2020 11:20
I have a GE Vivid I that sometimes ( like 3 times per week) shows this error message on boot. I will translate the message error from portuguese to English:
"ECG Trace currently disabled.
Restart system to reactivate ECG.
Continue --- Continue"
Then, the system crashes and i have to shut off pressing the OFF button for 3 seconds.
When the system turn on ok, it passes all the tests.
to rate and post a comment
ProemierEquipoMedico
re: GE Vivid I - ECG Trace Boot Error
December 10, 2020 10:28
May you have the signal board bad. It some times happen or can be software corrupted. If you need help contact us.