I have an issue with our system GE Vivid i system.
I've been setting up the connectivity to our Dicom/PACS server with success. When we do a new exam on a patient, it is sent to the spooler and transfers properly to our PACS. So far so good.
I am now trying to export the previous pateitns/exams and I cannot find the "export button". I have followed the training videos, the service and the user manual, but at the screen and moment they say to press the "export button", that button is just inexistant.
I have switched the language from our original one to english, just in case it was a programming issue. But it still isn't there.
Thank you for the suggestion!
I went to work in the settings of this echo for a bit, but the section "options" has the ones you mentionned as "permanent". So unfortunately, the problem must be somewhere else...
OMG I love you! Yes! That was what I needed to do!
Now I have a setting problems that I need to solve (totally unrelated), but after, It will be functionnal!
Therapeutic Technologies closeout auctions take place on DOTmed.com. Bid today on 538 lots of medical equipment. Inspections may be done by appointment. Hours are Monday through Friday 9am-4pm Contact chrisnuytens1@gmail.com to schedule.
Leonie Pilon
GE Vivid i no export button
November 21, 2022 10:30
Hello,
I have an issue with our system GE Vivid i system.
I've been setting up the connectivity to our Dicom/PACS server with success. When we do a new exam on a patient, it is sent to the spooler and transfers properly to our PACS. So far so good.
I am now trying to export the previous pateitns/exams and I cannot find the "export button". I have followed the training videos, the service and the user manual, but at the screen and moment they say to press the "export button", that button is just inexistant.
I have switched the language from our original one to english, just in case it was a programming issue. But it still isn't there.
Has anyone seen this before or has a solution?
to rate and post a comment
Nikolai Kuzhel
re: GE Vivid i no export button
November 28, 2022 10:03
May be you have locked options like DICOM, MPEGvue or EchoPAC and etc?
to rate and post a comment
Leonie Pilon
re: re: GE Vivid i no export button
November 29, 2022 10:56
Thank you for the suggestion!
I went to work in the settings of this echo for a bit, but the section "options" has the ones you mentionned as "permanent". So unfortunately, the problem must be somewhere else...
to rate and post a comment
DMYDPG
re: GE Vivid i no export button
December 12, 2022 10:24
In patient data field, change from int hdd/ dicom server to int hdd (second option from drop down menu) then the export button should show up.
to rate and post a comment
Leonie Pilon
re: re: GE Vivid i no export button
December 13, 2022 10:28
OMG I love you! Yes! That was what I needed to do!
Now I have a setting problems that I need to solve (totally unrelated), but after, It will be functionnal!
Thanks so much!