I have a problem with GE VMX .. error 432 ( Alim k2 time out ?? ).
All fuses are ok!
Thank you in advance

Log inor Register

to rate and post a comment

Robert Budach

Re: GE VMX error 430

October 09, 2011 01:26

Are you working on a GE AMX4? I do not know what a VMX is.

Log inor Register

to rate and post a comment

(24)
(6)

Jaime Diaz

Re: GE VMX error 430

October 09, 2011 08:03

Hi,

For VMX error 430 means:

If intermittent 430 error--capacitors not charged correctly.

This error may occur if the portable is prepped immediately after the unit is turned on. The commutating capacitors do not have enough time to charge. The problem could also be the measure circuit for the capcitors.

If it is an AMX4:
Error 430 during prep:Replace rotor board

Hope this information be useful.

Regards

1 person liked this
Log inor Register

to rate and post a comment

Josip Vujnovic

Re: GE VMX error 430

October 10, 2011 09:38

Thanks guys, solved!
By now I was in the office so I did not have access to service manual :-)
Yes, the error 430 is when the user presses the button too early for radiation ..This is not a defect but user errors.
In any case, thanks for your help

Best regards

Log inor Register

to rate and post a comment

(8)

Alrasana

re: Re: GE VMX error 430

May 21, 2016 03:56

Hello, I have error 430, on vmx plus , I measure all fuses are ok,
Hand switch , no one press exp while initialize
I took hand switch of from pcb , still the same
I kept it plugged in for sometime didn't help
Could please help
Thanks

Log inor Register

to rate and post a comment

Camilo Rodriguez

re: re: Re: GE VMX error 430

June 25, 2016 01:29

Do you still have this problem?

Log inor Register

to rate and post a comment

(8)

Alrasana

re: re: re: Re: GE VMX error 430

June 26, 2016 03:30

Yes sir

Log inor Register

to rate and post a comment

(8)

Alrasana

re: re: re: Re: GE VMX error 430

June 26, 2016 03:30

Yes sir , I notice if I press reset switch from CBU Board will work
And hear the noise of the inverter

Log inor Register

to rate and post a comment

Camilo Rodriguez

re: re: re: re: Re: GE VMX error 430

June 28, 2016 09:37

Have you measured voltage in the capacitor bank? (between F1 and F2 in the Converter Board.

Log inor Register

to rate and post a comment

(8)

Alrasana

re: re: re: re: re: Re: GE VMX error 430

July 04, 2016 10:35

I did replace CPU battery looks ok .? Is this the issue

Log inor Register

to rate and post a comment

(8)

Alrasana

re: re: re: re: re: Re: GE VMX error 430

July 05, 2016 09:47

Alrasanamed@gmail.com

Log inor Register

to rate and post a comment

stats
DOTmed text ad

MRI & CT Support Worldwide. Buy Parts Fast with LASO.

Comprehensive medical imaging support. We supply essential parts & maintenance for most OEMs, buy/sell quality used equipment, and offer mobile rentals globally. Contact us: info@lasoimaging.com

stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)