GE Voluson 730 PRO error BT_Interface

Ricardo A. Sanchez

GE Voluson 730 PRO error BT_Interface

August 26, 2011 08:59

Hi, I have a problem with a GE Voluson 730 PRO BT03, everytime I turned ON shows a message "BF_Interface FAULT" have you ever seen this before? I also noticed the "S/N" and the "Hardware ID" are not present, do you have any ide what can I do? let me know please, thanks

Log inor Register

to rate and post a comment

(46)
(9)
(1)

Victor Lupusor

Re: GE Voluson 730 PRO error BT_Interface

September 06, 2011 10:05

Perform a beamformer test to see if is defective,
However my guess is that it was replace a board in the system and you need to reprogram S/N. This is done with help from GE OLC.
Regards,
Victor

Log inor Register

to rate and post a comment

(37)
(4)

carlos mendez

Re: GE Voluson 730 PRO error BT_Interface

October 06, 2011 12:46

how can i contact ge olc, i have a similiar problem the systems options dont activate, and i put the key that comes whith the machine, how can i get them activated