Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

GE Voluson 730 pro Error on CRS and CPP

(11)
(3)

Darko Savic

GE Voluson 730 pro Error on CRS and CPP

July 22, 2015 08:27

Hello,
I am Darko from Serbia and I have GE Voluson 730pro BT08.

Can someone help me solve this problem...

when we turn on the Voluson software load correctly. At first glance everything works, but when you select eg. linear or vaginal probe appears the error: "Acoustic output or temperature limit could be exceeded. Error on the CRS and the CPP
in File .\I_Begrenzer\TxPowerToHW_Vx30.cpp:81"

Can someone help me to solve this problem.

Log inor Register

to rate and post a comment

(230)
(35)
(4)

Rey Gonzalez

re: GE Voluson 730 pro Error on CRS and CPP

July 22, 2015 04:32

The CRS is reporting the problem from the CPP, I would replace CPP if I have one available.
rey@allultrasoundservices.com

Log inor Register

to rate and post a comment

Ricardo A. Sanchez

re: GE Voluson 730 pro Error on CRS and CPP

July 23, 2015 01:58

Darko, yes I will say the same as Rey, check the CPP power supply, also, are all the fans running, when was the last time this system had a service done?

Log inor Register

to rate and post a comment

(11)
(3)

Darko Savic

re: re: GE Voluson 730 pro Error on CRS and CPP

July 23, 2015 08:07

Ok. The fans are cleaned and working correctly. the temperature is 26-27C. The system worked until December 2014. It was launched in April 2015, this error occurred. During the winter there was no heating in the room. Perhaps the cold affected the capacitors and other electronic components?

Does the CPP board can fix or do I have to buy another?
If someone can show me the ropes closer to a fault (eg. Replace the transistor xx or capacitor xx on cpp board ...), I'll buy some components and electrician will replace me.

Thank you for your patience.

Log inor Register

to rate and post a comment

(30)
(7)

akarsanov

re: GE Voluson 730 pro Error on CRS and CPP

July 23, 2015 08:07

1.) remove backup
(delete all files in folder
d:\UserSettings) and load
it (complete)
2.) reinstall UISAPP

Log inor Register

to rate and post a comment

(11)
(3)

Darko Savic

re: re: GE Voluson 730 pro Error on CRS and CPP

July 23, 2015 10:07

I replaced the new blank HDD and I installed the complete software. Error appeared again when I want to choose linear, vaginal or any other probe.

Log inor Register

to rate and post a comment

(11)
(3)

Darko Savic

re: GE Voluson 730 pro Error on CRS and CPP

November 04, 2015 01:47

Problem solved. Topics may be closed. Thank you all for your help.

Log inor Register

to rate and post a comment

(4)

Eng.FastoneOscar

re: re: GE Voluson 730 pro Error on CRS and CPP

December 24, 2015 06:33

Hi Darko Savic,
I can see that you solved the problem, But much better help to know what was the exact solution,among all ideals discussed above.
If you replaced the new pcb or you did this :
1.) remove backup
(delete all files in folder
d:\UserSettings) and load
it (complete)
2.) reinstall UISAPP
Regards Fastone