GE xray RT-1250DG problem

(18)
(1)

mario quetua

GE xray RT-1250DG problem

January 08, 2010 02:01

.when we select 200ma and much lower value
it appear error 03 but higher than 200ma its normal no error
appear...GE is the brand model is RT 1250dg..hopefully somebody will share the manual of this machine....thank you...

Log inor Register

to rate and post a comment

(244)
(19)

Staate Hayward

Re: GE xray RT-1250DG problem

January 12, 2010 10:35

I'm not framiliar with this model or system but did see on a foriegn web site that it appears to be a small doctors clinic / chiropractic type x-ray system. with the symptoms you described, it sounds like (1)the small filament may have blown out,(2)High Voltage connections at tube or transformer may be highly resistive or open, (3) Filament transformer or switch in the High Voltage Transformer, (4) Filament voltage control circuit board/driver?