GE XR/d Lightning table detector drive & Table SID Calibration

Larry Sprague

GE XR/d Lightning table detector drive & Table SID Calibration

November 13, 2011 01:07

Table CAM Motion Drive LED is Red, table detector won't "Home". Have replaced the CAM Motion Drive 3 times, replaced the table IF Board, the Table Logic Control board, and Table IO board. We are not sure of the DIP Switch setting for any of these items as someone else replaced some of these items. Need help with these switch settings. Assuming the present DIP Switch setting are correct, where else should we look to bring the CAM Motion Control on line?

Regarding DIP Switches, if you have the settings for the other boards we would like to compare to insure they are set correctly.

Also, having problems getting the table SID calibration completed. We understand that the CAM needs to be operational before the Table SID can be calibrated. This issue was existent before the CAM Motion Drive crashed. We were successful in calibrating the SID for the Chest Stand.

When looking at the part number for some of these boards there are two part numbers, why?

Thanks and appreciative for any help you might be
able to provide.

Larry

Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

Re: GE XR/d Lightning table detector drive & Table SID Calibrati

November 13, 2011 06:25

Do you have an error code?

Log inor Register

to rate and post a comment

Larry Sprague

Re: GE XR/d Lightning table detector drive & Table SID Calibrati

November 13, 2011 10:37

There are no error codes. Diagnostic software shows Pass or Fail and it does provide a short explanation that does not seem to provide enough information.

Thanks

Log inor Register

to rate and post a comment