Ethan MacIntosh
GE-DMR
November 20, 2017 05:34
Tube arc message at 49 kV on Mo track but not Rh track. Tube was replaced 7 months ago for same problem. Could this be an adjustment (job card)? Thank you. Ethan
to rate and post a comment
Maurice Dilick
re: GE-DMR
November 22, 2017 08:29
We can fix it, call Glen Harris at J & M Trading, Inc.
CRAIG WAMBURA
November 22, 2017 08:31
Use mammocom software to carry out HV and rotor diagnostics. If you do not have the software, I can share it. craig@telemed-technologies.com
Walter Gallegos
November 25, 2017 10:17
Run a HV test with and without HV cable cable. If both test pass claim tube warranty. I found several times this problem with non GE tubes, never with GE tubes; for records, please let me know gaine/insert brand and serial numbers.
Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net
Ethan MacIntosh
GE-DMR
November 20, 2017 05:34
Tube arc message at 49 kV on Mo track but not Rh track. Tube was replaced 7 months ago for same problem. Could this be an adjustment (job card)?
Thank you.
Ethan
to rate and post a comment
Maurice Dilick
re: GE-DMR
November 22, 2017 08:29
We can fix it, call Glen Harris at J & M Trading, Inc.
to rate and post a comment
CRAIG WAMBURA
re: GE-DMR
November 22, 2017 08:31
Use mammocom software to carry out HV and rotor diagnostics. If you do not have the software, I can share it.
craig@telemed-technologies.com
to rate and post a comment
Walter Gallegos
re: GE-DMR
November 25, 2017 10:17
Run a HV test with and without HV cable cable. If both test pass claim tube warranty.
I found several times this problem with non GE tubes, never with GE tubes; for records, please let me know gaine/insert brand and serial numbers.
to rate and post a comment