Over 500 Total Lots Up For Auction at One Location - NJ 06/15

GE_OEC error mA too high

(244)
(19)

Staate Hayward

GE_OEC error mA too high

July 11, 2007 02:54

We just replaced the x-ray tube on a 9800 after having tube arcs with "mA too low" messages. Now we are getiing "mA too high, Press any key to proced" when doing small anatomy ( ankle/leg ). Any ideas on what to do to correct this would be appreciated, 920-320-3339 shop, 905-3339 cell.

Log inor Register

to rate and post a comment

(244)
(19)

Staate Hayward

Re: GE_OEC error mA too high

July 11, 2007 05:28

I've since ran diags to get a couple of errors indicating the Fil_B sen failure, as well as charging_B battery. I'm not sure if this two problems and which bd.(s) to replace; Generator cntrl or Filament cntrl(driver).

Log inor Register

to rate and post a comment

(244)
(19)

Staate Hayward

Re: GE_OEC error mA too high

July 16, 2007 03:33

Thanks for the info, I ended up finding one of the priliminary null adjustments was off (outside of + - 1.5mV); after adjusting and running the mA cals again it stopped the errors at very low fluoro mA.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Multi-Vendor Imaging Service Partner

BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729

stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)