I have a problem in GECTE tubefan, it stop running during the exposure, and sometimes it run and stop and run and stop, all during the scan. the AC 108-115V, the neon is on.Rtation PCB were changed. The problem still there. any idea?
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
imeng
GECTE
October 18, 2014 02:59
HI Everybody,
I have a problem in GECTE tubefan, it stop running during the exposure, and sometimes it run and stop and run and stop, all during the scan. the AC 108-115V, the neon is on.Rtation PCB were changed. The problem still there. any idea?
staywell,
Imeng
to rate and post a comment
Edurado Madeiros
re: GECTE
October 20, 2014 10:51
Check the brushs and clean up slip ring too
to rate and post a comment
Lin Dong
re: GECTE
October 21, 2014 11:23
The fan is defective itself?