re: Genius HF generator from Villa - error message
September 16, 2017 04:26
NVRAM could be empty. Replace, put switch on processor board from normal to service then switch on. New NVRAM should be loaded from EPROM & all should be OK, if not you will have to program configuration again & calibrate filaments.
Comprehensive medical imaging support. We supply essential parts & maintenance for most OEMs, buy/sell quality used equipment, and offer mobile rentals globally. Contact us: info@lasoimaging.com
Adrian Mihaita
Genius HF generator from Villa - error message
September 15, 2017 09:19
Hello,
At startup, the console desk show the message " Error, call service team ". Do you have any ideea ?
On CCB board the red LED's DL16,17,18 are on.
Thank you !
to rate and post a comment
LUIS ERNESTO OROZCO
re: Genius HF generator from Villa - error message
September 16, 2017 04:25
Hi Adrian i can help you, email me saisco.soluciones@gmail.com
to rate and post a comment
Lister
re: Genius HF generator from Villa - error message
September 16, 2017 04:26
NVRAM could be empty. Replace, put switch on processor board from normal to service then switch on. New NVRAM should be loaded from EPROM & all should be OK, if not you will have to program configuration again & calibrate filaments.
to rate and post a comment
Adrian Mihaita
re: re: Genius HF generator from Villa - error message
September 19, 2017 09:00
You suggested the right solution. Everything is OK.
Thank you !
to rate and post a comment
Ahmad Amache
re: re: Genius HF generator from Villa - error message
May 13, 2025 05:37
Hello , can you help me with same issue
to rate and post a comment
DMEDS
re: Genius HF generator from Villa - error message
September 24, 2017 10:12
Dear Sir,
I have all manuals for this system, if you need it contact me,
magmesme@gmail.com
Regards
to rate and post a comment