Hello everybody, I have a problem with a Philips Multidiagnost eleva, I install the system but I have problema with The Geo movements, it's appear like collision, but there is not, so I can´t calibrate the movements, either the up and down table move can´t do it in service or operator mode, I don´t have all the documentation to can detect what the problem is, so if somebody can help with some info, please let me know to lobsang_ortiz@hotmail.com
Regards
I experienced similar problem on easy diagnost system, i am not sure elevas gantry is the same or not.
There is a gravity safety switch on the back of the gantry which is maybe activated. It can be released from the balance weight cavity from the leg side.
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
Lobsang Ortiz
Geo problem in Philips multidiagnost Eleva
April 23, 2019 10:15
Hello everybody, I have a problem with a Philips Multidiagnost eleva, I install the system but I have problema with The Geo movements, it's appear like collision, but there is not, so I can´t calibrate the movements, either the up and down table move can´t do it in service or operator mode, I don´t have all the documentation to can detect what the problem is, so if somebody can help with some info, please let me know to lobsang_ortiz@hotmail.com
Regards
to rate and post a comment
Tibor Ivanszky
re: Geo problem in Philips multidiagnost Eleva
April 24, 2019 10:23
I experienced similar problem on easy diagnost system, i am not sure elevas gantry is the same or not.
There is a gravity safety switch on the back of the gantry which is maybe activated. It can be released from the balance weight cavity from the leg side.
to rate and post a comment