gradient problem g.e. 9x

Tags: MRI


roman solano jesus antonio

gradient problem g.e. 9x

November 26, 2014 08:17


Hello good day I have this problem:
the gp reported a gradient amplifier z axis internal wiring fault or control undervoltage

the gp reported a gradient amplifier y axis internal wiring fault or control undervoltage

error number:
2244573

automatically makes recovery

help me thanks

Log inor Register

to rate and post a comment

(102)
(39)
(1)

Konstantin Kuznetsov

re: gradient problem g.e. 9x

November 27, 2014 09:41

LX ACGD?
1. Check the 208V power and cable (maybe not completely seated at the Axis SGA). Check a gradient amplifiers, adjust them.
2. For LX EchoSpeed or TwinSpeed systems, with ACGD running ASP2 or 9.x software: Intermittent SGA Internal Wiring Fault and Internal Power Supply Undervoltage. May be associated with oblique imaging, but also seen on sagital FSE scans using a knee coil and very small FOV. The fault may be somewhat repeatable with a given sequence, but will clear up if the FOV is increased. May be the real root cause of this problem with the GP:
There is one IC to be replaced in the GP. It is an IC to be populated at U108 on the GP, marked "gg513"

Log inor Register

to rate and post a comment

(4)
(1)

Yasir Raza

re: gradient problem g.e. 9x

November 27, 2014 09:41

1st of all Try to replace the CPU boards with any other amplifier....and check the results

Log inor Register

to rate and post a comment

mahmoud abdelhaddy

re: gradient problem g.e. 9x

November 27, 2014 09:42

it is may be gradient power supply
you may swap between gradient to check if the error disappered or not for any help
en.mahmoud.abdelhady@gmail.com

Log inor Register

to rate and post a comment

Joe Valdez

re: gradient problem g.e. 9x

January 10, 2015 06:09

Look at SGA-PS front LEDs. If BUS OV LED lit, check PDU 420V output. If OV X,Y,Z LED lit, swap military connectors to isolate SGA vs. SGA-PS. Power cycle ACGD POWER breaker on PDU to recover.