Dear
I have trouble with Hispeed QXi. When I warm up and scan, on screen show: Scan have stopped and log file show:
SYSTEM POWER CONTROL: HVDC Control
I/O board pilot relays are closed. PDU relay not energized.
BCSTAT:=0 HVDC Rail Voltage = 206V
What sleepers of pdu is installed? CPDU or NGPDU?
-check the input voltage to the PDU
-check all voltages.
individually check-80A, 600V, semiconductor fuse.
check-110A, 1600V diode bridge
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
Tuy Do Duy
Have problem with Hispeed QXi
January 19, 2013 11:28
Dear
I have trouble with Hispeed QXi. When I warm up and scan, on screen show: Scan have stopped and log file show:
SYSTEM POWER CONTROL: HVDC Control
I/O board pilot relays are closed. PDU relay not energized.
BCSTAT:=0 HVDC Rail Voltage = 206V
Who have solutions?
to rate and post a comment
roger dussart
re: Have problem with Hispeed QXi
January 19, 2013 06:01
For me, the problem is in PDU check fuses and breaker
to rate and post a comment
Francesco Palmieri
re: Have problem with Hispeed QXi
January 20, 2013 01:35
What sleepers of pdu is installed? CPDU or NGPDU?
-check the input voltage to the PDU
-check all voltages.
individually check-80A, 600V, semiconductor fuse.
check-110A, 1600V diode bridge
to rate and post a comment
Luis Santos
re: Have problem with Hispeed QXi
January 29, 2013 09:10
Check the caps next to the rectifier bridge, use your multimeter on capacitance. I bet one is bad.