Have problem with Hispeed QXi

Tags: CT, GE

(91)
(14)

Tuy Do Duy

Have problem with Hispeed QXi

January 19, 2013 11:28

Dear
I have trouble with Hispeed QXi. When I warm up and scan, on screen show: Scan have stopped and log file show:
SYSTEM POWER CONTROL: HVDC Control
I/O board pilot relays are closed. PDU relay not energized.
BCSTAT:=0 HVDC Rail Voltage = 206V

Who have solutions?

Log inor Register

to rate and post a comment

(87)
(1)

roger dussart

re: Have problem with Hispeed QXi

January 19, 2013 06:01

For me, the problem is in PDU check fuses and breaker

Log inor Register

to rate and post a comment

Francesco Palmieri

re: Have problem with Hispeed QXi

January 20, 2013 01:35

What sleepers of pdu is installed? CPDU or NGPDU?
-check the input voltage to the PDU
-check all voltages.
individually check-80A, 600V, semiconductor fuse.
check-110A, 1600V diode bridge

Log inor Register

to rate and post a comment

(12)
(3)

Luis Santos

re: Have problem with Hispeed QXi

January 29, 2013 09:10

Check the caps next to the rectifier bridge, use your multimeter on capacitance. I bet one is bad.