Dear Friends
I have a problem with one HDI 5000 when turn ON I see in back of APM module and the 4 led only one is turn green and the other 3 in red, later appear in blue screen error 89, in the past day in cold time the system is turn on fine and now appear again this error 89 (to 25.6 degres), we delete this error and again appear when turn on.
I believe is a problem with the APM module with some component that overheat, but we not have the service manual of that module, some body fix this problem of overtemp in some component?
Thank you all

1 person liked this
Log inor Register

to rate and post a comment

(13)
(4)

Skip Kirchhoff

re: HDI 5000 Error 89,,help.

November 17, 2012 01:01

Remove APM. Remove Hard drive. Connect to another computer. Run CHKDSK on it. replace everything.

Log inor Register

to rate and post a comment

(35)
(2)
(3)

Robert Celmer

re: HDI 5000 Error 89,,help.

November 17, 2012 12:58

Parts@nwmedicalequipment.com easy fix

Log inor Register

to rate and post a comment

Ricardo A. Sanchez

re: HDI 5000 Error 89,,help.

November 17, 2012 12:59

check the DVS, you have a connection problem or is not working and the system can't communicate with the DVS

Log inor Register

to rate and post a comment

(6)
(2)
(1)

Khaled Abed

re: HDI 5000 Error 89,,help.

November 17, 2012 12:59

Hi,

It happens to one of my HDI5000 machines (newer skin) and what I do usually is:
- reseat both cables on the back of the machine (that connect APM to the machine)
- reboot

Always the problem disappears but comes back after 2-3 months and I do the same thing again.

It's most likely a bad contact on those connctors not a heat related issue.

Sorry I'm not in front of the machine now but would give you more details on Monday if needed. Let me know If you know which two connectors I'm talking about.

Good luck!

Khaled

Log inor Register

to rate and post a comment

(230)
(35)
(4)

Rey Gonzalez

re: HDI 5000 Error 89,,help.

November 18, 2012 02:02

you havre to check if the APm or DVS is booting. for help contact me rey@allultrasoundservices.com

Log inor Register

to rate and post a comment

(43)
(3)
(4)

Rodolfo Reyes Ingeniero

re: re: HDI 5000 Error 89,,help.

December 26, 2012 06:00

Thank you, the problem was solved only reconnect the boards inside the module.thankyou

Log inor Register

to rate and post a comment

(43)
(3)
(4)

Rodolfo Reyes Ingeniero

re: HDI 5000 Error 89,,help.

December 26, 2012 06:00

Dear Friends thanks to all for your recommendation , we open the module, and disassemble the boards, return again to connect to motherboard and work fine , the problem is when transport the boards maybe disconnect and for that reason not work and appear the error 89

Log inor Register

to rate and post a comment

stats
DOTmed text ad

BK and other Urological Systems available. BK Probe repair available!!!

Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net

DOTmed Certified stats
advertisement

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)