HDI 5000 errors ## 0021,0019,0084,0089

November 11, 2006 07:58

Dear Colleagues, We have a problem with Ultrasound HDI-5000 SonoCT. It can not detect its scanheads (initialization takes very long time and ends with an error) Working without scanheads it displays error messages with errors ## 0084 (the most frequently), 0089, 0019 and 0021 (sometimes). What does it mean? What does the problem connect with? Help us, please. We do need to solve the problem. Thank you. Alex MEDIVIT Ltd. Mailto:reutaav@inbox.ru

Log inor Register

to rate and post a comment

Jim Carr

Re: HDI 5000 errors ## 0021,0019,0084,0089

November 13, 2006 11:14

Alex,
0084 and 0089 point to problems with the DVS (computer). 0021 and 0019 are due to "task crash" and 0021 is sometimes associated with a loss of network connection. Check the ethernet connections and the power cable to the DVS.

Jim Carr, Sonora Medical Systems

Log inor Register

to rate and post a comment

(26)
(3)

Alexandr Reuta

Re: HDI 5000 errors ## 0021,0019,0084,0089

November 16, 2006 03:20

Dear Jim Carr,

Thank you very much for your answer. I hope it'll help us to reduce the problem.
Best Wishes

Alex Reuta, Medivit Ltd.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

BK and other Urological Systems available. BK Probe repair available!!!

Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net

DOTmed Certified stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)