manuel valdez
hdi 5000 problems
July 30, 2013 06:40
hello all
i have a problem with a HDI 5000 system
i run the diagnost in service mode and result varius errors
error: subsystem=feature-diags, Errno= 0x20001124 and 0x200011ca,0x20000102
when i finished the test reboot the system and the next time that the equipment turn on displayed the next:
ALERT
An Error has been detected
please record the following information and call ATL technical Supoport
date: 07/30/2013 time: 12:09:45
Diagnost number: 0017
and
System Error
there is a problem with the system.
Reselect a scanhead. ifthere is stiill a problem, call your service representative and report the error .
ok
i dont have the service manual some body can help me to fix the problem?? in advance thanks
to rate and post a comment
Ricardo A. Sanchez
re: hdi 5000 problems
July 31, 2013 08:39
contact me ricardo@ultrasolutions.com
to rate and post a comment
Miguel Flores
re: hdi 5000 problems
August 01, 2013 08:28
Hello.
I'm from Mexico too. qro.
I know very well this sistem.
I have the service manual.
If you want i can help you.
contact me biozig@gmail.com
to rate and post a comment
Ricardo A. Sanchez
re: hdi 5000 problems
August 01, 2013 03:24
0017 bUtilStatus.c Slot configuration error on bootup. Review the previous error messages and
the suspected slot and board will be identified. Either the configuration was
read incorrectly which would be a board failure or the board was not seated
correctly. Another possibility is that the data files that describe how the
center plane is configured (brdList.xml, ctrP96.xml, sltP96.xml) are invalid or
corrupt
to rate and post a comment
ricardo scacchi
re: hdi 5000 problems
August 06, 2013 08:23
meaning of error code 0017
bUtilStatus.c Slot configuration error on bootup. Review the previous error messages and
the suspected slot and board will be identified. Either the configuration was
read incorrectly which would be a board failure or the board was not seated
correctly. Another possibility is that the data files that describe how the
center plane is configured (brdList.xml, ctrP96.xml, sltP96.xml) are invalid or
corrupt.
But in my opinion first of all , you must check the power supply values, I think there something is wrong , then check the boards location using the label at the inner side of the plastic front and rear cover
All the best Ricardo