Over 150 Missouri Auctions End Today - Bid Now

hdi 5000

(93)
(9)

kenan celik

hdi 5000

May 23, 2007 01:53

hdi 5000 color doppler machine giving the beep sound when i push the power(some times)and it is not work,i think this is a power problem but i am not sure,is this a power supply problem or electrical power problem.because machine normal working other place.same body help me?

Log inor Register

to rate and post a comment

Jason Munson

Re: hdi 5000

May 24, 2007 09:30

Hi there!

This sytem has some good diagnostic tools for troubleshooting.

With the main power switch (breaker) ON, but the ON/Standby switch OFF, please verify on the ACIM board LEDs #6 and #7 are continuously lit. LED #6 indicates power is available to the HVDC transformer after the fuse, and LED #7 indicates HVDC is available to the PSM.

The beep you are hearing may indicate HVDC is not available to the PSM, preventing the system from starting.

If you have other questions, please post to this forum.

Respectfully,

Jason Munson
Echoserve, LLC
877-386-3246

Log inor Register

to rate and post a comment

(93)
(9)

kenan celik

Re: hdi 5000

May 24, 2007 10:09

dear sir
i did chek the leds and hvdc is not available.
but electrical voltage 220 volt in the building.
i think this a electrical to ground problem.
thank your to answer.