HDI 5000

TS

HDI 5000

May 11, 2017 03:44

I need some help with my 5DI 5000. It seems to not penetrate past about 5cm. I can see some image below 5cm but it is faint. It doesn't matter what probe I am using(I have more than on variable Freq.) Adjusting the gain increases sound on the portion of tissues closest to the probe head but doesn't seem to change much past 5cm. Any thoughts?

Log inor Register

to rate and post a comment

(119)
(57)
(6)

Jim Carr

re: HDI 5000

May 12, 2017 07:45

Assuming the Output control setting has not been accidentally reduced, I would check the transmitter voltage power supply LED, the top one on the PSM. The transmitter voltages should be +/- 115vdc at maximum. (Use an adult abdominal or cardiac preset and make sure acoustic Output is at maximum.)

Log inor Register

to rate and post a comment

TS

re: re: HDI 5000

May 13, 2017 09:08

I don't have an adult abdominal setting. I purchased this machine with the hope of using it on dogs/cats(I am a veterinarian). I will check the voltage on the psm.

Log inor Register

to rate and post a comment

(77)
(18)

Brian Eberling

re: HDI 5000

May 12, 2017 07:46

Have you ran the diagnostics on the system?

Log inor Register

to rate and post a comment

Tony Sheet

re: re: HDI 5000

May 13, 2017 09:12

No. Could you help me do this?

Log inor Register

to rate and post a comment