Over 500 Total Lots Up For Auction at One Location - NJ 06/15

HDI3000 power on error

(432)
(83)
(3)

juan pablo amado

HDI3000 power on error

December 23, 2007 06:56

At power on, leds in input power module lit in this order :5,4,3,2,1 then 4,3,2,1. Never 6 and 7 lit on. Then power down and 1,2 remain flashig.

Wath is the cause?

Log inor Register

to rate and post a comment

Jim Carr

Re: HDI3000 power on error

December 26, 2007 01:07

Sounds like you have lost the high voltage since both LEDs 6 and 7 should be on. ACIM LED 6 indicates power to HVDC after internal fuse F1 and ACIM LED 7 indicates +170VDC from rectifier to the MPS and APS. The HVDC is generated on the ACIM.

Note: LEDs 1 to 5 should scroll sequentially and LEDs 6 and 7 be steady on in normal operation.

Log inor Register

to rate and post a comment

(93)
(9)

kenan celik

Re: HDI3000 power on error

December 26, 2007 01:33

hi.
i gest your problem is a power supply error.
this machine have 3 power boards:master,analog,dıjıtal
yo must control the master power LED's
it can be 5 volt digital power supply problem
if flashing led (5 volt dıgıtal power)on master power bord,
it can be 5 volt digital power supply problem

Log inor Register

to rate and post a comment

(432)
(83)
(3)

juan pablo amado

Re: HDI3000 power on error

January 17, 2008 08:06

Leds 6 and 7 are ON ( vision error) always.
In master PS
At self power off,
SCIP ( DS1) flashing slow
Analog (DS2) and Digital (DS3) flashing fast.

Do you have schematics of this board?

Log inor Register

to rate and post a comment

(93)
(9)

kenan celik

Re: HDI3000 power on error

January 18, 2008 03:23

you don't need schematics of this board!
it is a DC/DC converter error probable(vicor) on the dijital power supply or other power supply.
you can change them.


Log inor Register