help MPX 100 GE

Julian Aguilera

help MPX 100 GE

November 04, 2006 12:46

I was instaled MPX 100 with RFX, XT, FL 300, but I have trouble with micro computer program. I need some help venexray01@yahoo.com

Log inor Register

to rate and post a comment

Ralph Childs

Re: help MPX 100 GE

November 04, 2006 10:13

Julian,

I have all of the books for these units. However, they are really tough to work on. They were bad when new and much worse now that they are 20 years old.

Try to get me more info and I will send you as much help as possible.

Scooter

Log inor Register

to rate and post a comment

Julian Aguilera

Re: help MPX 100 GE

November 04, 2006 01:31

I have MPX Vol 1 & 2. I was make vol 1 charter 3. step 4. in this moment I have activated leds on MG2A6: MMR ; WAIT ; RUN; FTCH at same time. How can I checking the corret program for work?. where are the problem?. I can't made automatic calib.

thanks a lot.

Log inor Register

to rate and post a comment

(6)

(2)

Nimrod McDade

Re: help MPX 100 GE

November 04, 2006 06:08

Are your memory boards corrupt?
Are your power supplies good?
Do you know how to load your memory boards?

Log inor Register

to rate and post a comment

Julian Aguilera

Re: help MPX 100 GE

November 04, 2006 10:33

Hi Nimrod,

yes, may be that memory boards are corrupt. but how I can check this. how I can compare whith another good.
I was replaced batteries, ten years old. I know how to load memory in boards, adress and data.

thanks

Log inor Register

to rate and post a comment

Ralph Childs

Re: help MPX 100 GE

November 04, 2006 11:13

Julian,

The uP is frozen. It could be any number of thinngs. First check al of teh DC power supplies. Then remove all of teh non-necessary pcb's and try tthe cpu alone. If it still hangs then either your software is wrecked or the processor has died.

Log inor Register

to rate and post a comment

Julian Aguilera

Re: help MPX 100 GE

November 08, 2006 05:42

Hi Ralph,

I was check all LVDC power supplies and they are good.
For calib parameter I was data load manual in memory board for tube 1. now I can get exposure. but when an exposure was finished the alarm go on. Only I can made another exposure if "Reset" on MG2A6A47 or the console is turned off. what hapen with it?? only I have 20Kvp aprox diferent of the selecting.

When I select vertical bucky (botton 4)don't begin exposure.


The "Leds ON" on panel MG2A6 are: MMR, WAIT,RUN,FTCH. at same time.
let me know if you have any idea or need more info.

thanks





Log inor Register

to rate and post a comment