am a service Biomedical Engineer in Ghana working on Phillips Mx8000 which gives artifact in all protocol i need help if any one can help contact me on nkansah2g5@yahoo.com i will be very glad to hear from ou my cologues
Hi. I have a quad. and after 5 or 6 procedures (abdominal or brains), appears a ring artifcat in the middle or in other moment in an upper position. The temp is 23 dg, ok, and humid 32 ok.
What could it be?.thanks in advanced
try fisrt to do a reconsruction of a good raw data image fromyour console, with this you will check if the problem is in the reconstruction box or the problem is in the gantry, how old is the tube????? have you checked that the filements are not shifting????, we need more information....
Go to ring segment menu or bad detector ring id... replace the detector and do air cal and phantom cal head and body... delete the table in bad segment... try to see bad segment menu click auto then delete previous then add the new table.. for mx16 idt system same procedure
GBS offers service, repair, rentals & spare parts for your ventilators. With over 40 years of experience our certified & factory-trained biomedical technicians refurbish all our used ventilators. Call today 800.558.9449
samuel nkansah
help on Mx8000
November 15, 2011 10:05
am a service Biomedical Engineer in Ghana working on Phillips Mx8000 which gives artifact in all protocol i need help if any one can help contact me on nkansah2g5@yahoo.com i will be very glad to hear from ou my cologues
to rate and post a comment
Fred Faour
Re: help on Mx8000
November 18, 2011 10:21
What kinda off MX800 You have ? 2 Slice ? 4 Slice ?
to rate and post a comment
Firas Afifi
Re: help on Mx8000
November 21, 2011 07:48
In the service menu, try to find the auto ring ID
to rate and post a comment
Primco
re: Re: help on Mx8000
May 23, 2013 08:22
Hi. I have a quad. and after 5 or 6 procedures (abdominal or brains), appears a ring artifcat in the middle or in other moment in an upper position. The temp is 23 dg, ok, and humid 32 ok.
What could it be?.thanks in advanced
to rate and post a comment
chalie alvarez
re: re: Re: help on Mx8000
May 23, 2013 10:36
try fisrt to do a reconsruction of a good raw data image fromyour console, with this you will check if the problem is in the reconstruction box or the problem is in the gantry, how old is the tube????? have you checked that the filements are not shifting????, we need more information....
to rate and post a comment
Primco
re: re: re: Re: help on Mx8000
May 24, 2013 10:16
Hi. I have done the reconstruction of raw data and appears fine. The tube has 4 years used. How can i check if the filaments are shifting?
to rate and post a comment
junn refol
re: re: re: re: Re: help on Mx8000
September 11, 2014 12:56
Go to ring segment menu or bad detector ring id... replace the detector and do air cal and phantom cal head and body... delete the table in bad segment... try to see bad segment menu click auto then delete previous then add the new table.. for mx16 idt system same procedure
to rate and post a comment