help on Mx8000

samuel nkansah

help on Mx8000

November 15, 2011 10:05

am a service Biomedical Engineer in Ghana working on Phillips Mx8000 which gives artifact in all protocol i need help if any one can help contact me on nkansah2g5@yahoo.com i will be very glad to hear from ou my cologues

Log inor Register

to rate and post a comment

(3)

Fred Faour

Re: help on Mx8000

November 18, 2011 10:21

What kinda off MX800 You have ? 2 Slice ? 4 Slice ?

Log inor Register

to rate and post a comment

(20)

Firas Afifi

Re: help on Mx8000

November 21, 2011 07:48

In the service menu, try to find the auto ring ID

Log inor Register

to rate and post a comment

(11)

Primco

re: Re: help on Mx8000

May 23, 2013 08:22

Hi. I have a quad. and after 5 or 6 procedures (abdominal or brains), appears a ring artifcat in the middle or in other moment in an upper position. The temp is 23 dg, ok, and humid 32 ok.
What could it be?.thanks in advanced

Log inor Register

to rate and post a comment

(45)
(1)

chalie alvarez

re: re: Re: help on Mx8000

May 23, 2013 10:36

try fisrt to do a reconsruction of a good raw data image fromyour console, with this you will check if the problem is in the reconstruction box or the problem is in the gantry, how old is the tube????? have you checked that the filements are not shifting????, we need more information....

Log inor Register

to rate and post a comment

(11)

Primco

re: re: re: Re: help on Mx8000

May 24, 2013 10:16

Hi. I have done the reconstruction of raw data and appears fine. The tube has 4 years used. How can i check if the filaments are shifting?

Log inor Register

to rate and post a comment

junn refol

re: re: re: re: Re: help on Mx8000

September 11, 2014 12:56

Go to ring segment menu or bad detector ring id... replace the detector and do air cal and phantom cal head and body... delete the table in bad segment... try to see bad segment menu click auto then delete previous then add the new table.. for mx16 idt system same procedure

Log inor Register